¿ªÔÆÌåÓý

Date

Re: Frequency References

 

Roy, with one difference... Time and volt nuts chase quantifiable metrics with measureable and definable absolute values using standardized nomenclature and accepted methodology. Audiophools measure nebulous metrics with values such as "warmth", "brightness", and my favorite, "presence", while throwing out the factual formulas and ideas like ohms law and human ear response because they love brand X better than Y. Also, in the audiophool world, quality is always directly proportional to the cost if the device. ?


Re: HP 3325

 

¿ªÔÆÌåÓý

The part numbers you found are assemblies, which probably means the transistors with the leads attached. The transistor part numbers are:

1853-0450 MJE371K (selected)
1854-0800 (not listed in the widely available cross reference)
1853-0453 TIP36A

1853- parts are PNP transistors?
1854- are NPN



On Jun 20, 2024, at 15:03, Joe White <skinnershorse@...> wrote:

?These are from the power supply board of a 3325.? Per OSM they are p/n 03325-66901, 03325-66902, 03325-66903 (left to right), described as "XSTR ASSEMBLY".? The manufacturer is listed as HP.? I have not found the part numbers in any cross-reference lists and I have been unable to identify the parts by their markings.? Any help on identifying the parts and/or replacements is greatly appreciated.
--
Joe White
KW4YW
<3325 xstr.JPG>


Re: HP 3325

 

Hi Joe

HP part #

1853-0450 =?MJE371K (not sure)
1854-0800 -> can't find this number
1853-0453 = TIP36A

Best?
Hans


Den torsdag den 20. juni 2024 kl. 22.03.35 CEST skrev Joe White <skinnershorse@...>:


These are from the power supply board of a 3325.? Per OSM they are p/n 03325-66901, 03325-66902, 03325-66903 (left to right), described as "XSTR ASSEMBLY".? The manufacturer is listed as HP.? I have not found the part numbers in any cross-reference lists and I have been unable to identify the parts by their markings.? Any help on identifying the parts and/or replacements is greatly appreciated.
--
Joe White
KW4YW


HP 3325

 

These are from the power supply board of a 3325.? Per OSM they are p/n 03325-66901, 03325-66902, 03325-66903 (left to right), described as "XSTR ASSEMBLY".? The manufacturer is listed as HP.? I have not found the part numbers in any cross-reference lists and I have been unable to identify the parts by their markings.? Any help on identifying the parts and/or replacements is greatly appreciated.
--
Joe White
KW4YW


Re: Frequency References

 

I'll attach some block diagrams of the various setups when I get home tonight.? Can't draw stuff up in any reasonable amount of time on my phone here at work.? I'll show the 50 ohm 10 MHz distribution method we have at work, the method I use in my home lab, and the 75 ohm baseband analog video distribution method.? ? ?Jim


On Thu, Jun 20, 2024 at 9:16 AM, Jinxie via groups.io
<paul666@...> wrote:
So having read all the replies so far - for which many thanks - is it the same reason that all the PLLs within my 8566B spectrum analyzer are not only locked themselves but inter-locked with each other as well? So to ensure everything's in phase with everything else?


Re: Frequency References

 

So having read all the replies so far - for which many thanks - is it the same reason that all the PLLs within my 8566B spectrum analyzer are not only locked themselves but inter-locked with each other as well? So to ensure everything's in phase with everything else?


Re: Agilent E7495A Firmware Above A.3.10

 

Good news! I'm glad it worked and I could help.

73!
Razvan

On 20/06/2024 16:19, Charles Edmonds via groups.io wrote:
Hi Razvan

thank you, it worked OK.

I tried it first with a 2GB compact flash card but that didn't work so I
tried a 256MB card and it worked OK so presumably it does not recognise
such a large card.

Thanks again.

regards
Charles Edmonds VK3CLE
Mob: 0418309921
charles.edmonds@...


Re: Frequency References

 

Ed: The Leo B GPS (and most GPS/TCXO etc) have Zo 50 Ohms.
We have a stock of vintage BNC 2X, 4X, 6X 50 Ohm wideband splitters.
These are used often for video , so perfect BW for 5..10 MHz.
Running the GPS at max output, the losses in the splitters (use transformers) still give > 500 Mv at the loads.

Our setup: Leo B 10 MHz>>Mini Circuits 2X>>Mini Circuits 4X, 6x etc.
We drive 6 to 10 devices: Counters, spec/network analyzers, scopes.

Min-Circuits are good cost/availablilty.


Jon


Re: Are HP 70900A and HP 70900B interchangeable ?

 


Apologies? typo error:
Not 70902,
.... should read 70900a > 70900b


Re: Agilent E7495A Firmware Above A.3.10

 

Hi Razvan

thank you, it worked OK.

I tried it first with a 2GB compact flash card but that didn't work so I tried a 256MB card and it worked OK so presumably it does not recognise such a large card.

Thanks again.

regards
Charles Edmonds VK3CLE
Mob: 0418309921
charles.edmonds@...


Re: Are HP 70900A and HP 70900B interchangeable ?

 


hi Ron.

For additional info:

See:
the Hp70k thread on the eevBlog:



The Hp 70902b FW can be upgraded to a newer version.
There *might* be an (unreasonable) way to mod a 70902a > to a "b" version,
.... but would be an extensive rabbit hole !

See:
Groups.io files section /g/hp70000/files
and KO4BB.



Re: Frequency References

 

I assume that you distribute the reference through a resistive splitter to each device.
The output of the LEO is a square wave and long cable lengths can cause waveform distortion due to reflections, especially for those units tapped off along the line.

ed


Re: Agilent E7495A Firmware Above A.3.10

 

Hello Charles,

I uploaded the last firmware released which was A06.25 and also the
Measurement Guide and Service Guide manuals.

/g/HP-Agilent-Keysight-equipment/files/All%20HP,%20Agilent%20and%20Keysight%20instruments%20in%20folders%20by%20part%20numbers/7000%20to%207999/E7495A_B

Firmware is the same for both A/B versions.

Regards,
Razvan

On 20/06/2024 09:22, Charles Edmonds via groups.io wrote:
Hi,

I have an E7495A with F/W A.03.10-0.0.0.0. I can't find anything more
recent on the Keysight web site or anywhere else. I particularly want to
get at least A.03.20 as that adds a Sample Marker so should be better
for measuring noise levels.

Hoping someone can help please.

regards
Charles Edmonds
(Melbourne, Australia)
_._,_._,_


Agilent E7495A Firmware Above A.3.10

 

Hi,

I have an E7495A with F/W A.03.10-0.0.0.0. I can't find anything more recent on the Keysight web site or anywhere else. I particularly want to get at least A.03.20 as that adds a Sample Marker so should be better for measuring noise levels.

Hoping someone can help please.

regards
Charles Edmonds
(Melbourne, Australia)??


Re: Frequency References

 

We used the excellent Leo Bodnar GPS for years,



USB and Windows interface allow selection of F and V, 800 hZ..800 MHz, out to 32 dbm

We set the F to 10.000 MHz and output to MAX.

We distribute the ref to all our TEK/HP/other via Mini circuits splitters, one Leo B GPS drives 5..10 50 Ohm ref inputs.

Enjoy

Jon


Re: Frequency References

 

Hi there!

I'm not a time nut, but almost. I like my equipment to go at the same rythm, so I spent about €200 on two boxes: a GPSDO and an 8-output distribution amplifier and I am very happy with the result. My frequency measurements are precise and synchronized enough. Those were bought at eBay, from China, you know, that green and aluminum small boxes.

Regards,
Ruben


File /Winchester-Rack-and-Panel-Brochure.pdf uploaded #file-notice

Group Notification
 

The following files and folders have been uploaded to the Files area of the [email protected] group.

By: RubenRubio <rubenrb2019@...>

Description:
Cataloge for Winchester Electronics Connectors present in 8660 synthesizer system mainframe and plugins


Re: Frequency References

 

On June 19, 2024 10:10:57 PM "Dave McGuire" <mcguire@...> wrote:
On June 19, 2024 9:59:33 PM "Roy Thistle" <roy.thistle@...> wrote:
Time Nut
'time nut' ... 'volt nut'... or 'audiophile'... IMO... they sometimes make it harder for other 'nuts' to get the resources they need to actually 'do' stuff.
The difference bei6, of course, that the former two actually know something about technology.
(err, "being")

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Frequency References

 

On June 19, 2024 9:59:33 PM "Roy Thistle" <roy.thistle@...> wrote:
Time Nut
'time nut' ... 'volt nut'... or 'audiophile'... IMO... they sometimes make it harder for other 'nuts' to get the resources they need to actually 'do' stuff.
The difference bei6, of course, that the former two actually know something about technology.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Frequency References

 

On Wed, Jun 19, 2024 at 09:03 AM, Matt Harris wrote:
Time Nut
'time nut' ... 'volt nut'... or 'audiophile'... IMO... they sometimes make it harder for other 'nuts' to get the resources they need to actually 'do' stuff.