You can get the connectors from binder which is preferred if you are in Europe ( RS Components , buerklin, farnell )? or you can buy them from amphenol !
A few words of wisdom though based on experience :
First of use connectors that are gold plated if you use other plating options you can end up with issues at low power levels ( noise ? thermal voltages ???? ) the second thing
is that you have to follow the original stackup of the cable , the agilent cable uses 12 wires but 2 wires are shielded separately and you have to keep it that way to avoid problems !
Back when I build up cables what I did is to order a 100ft 11730 cable from agilent which cost me about $1200 , I cut the cable in 3ft sections ordered C091 Amphenol connectors
and then used them to build the cables !
?
https://www.mouser.com/datasheet/2/18/C09131C0122002-1285663.pdf
?
?
toggle quoted message
Show quoted text
From: HP-Agilent-Keysight-equipment@groups.io <HP-Agilent-Keysight-equipment@groups.io>
On Behalf Of Paul Bicknell via groups.io
Sent: Thursday, May 19, 2022 3:10 PM
To: HP-Agilent-Keysight-equipment@groups.io
Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
?
The connecters are available, Binder make 3 types and they did cost ?10 , ? 13 ?and ?16? each
?
Regarding Diode sensors do not forget the 430 meter was made for Diode sensors? also I believe the early 431 A had a switch position for Diode
?
Personally looking for a modern meter ( not expensive ) for wave guide diode detectors but calibrated in dB
Paul
?
From:
HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io]
On Behalf Of Alexandre Souza
Sent: 19 May 2022 20:49
To: HP-Agilent-Keysight-equipment@groups.io
Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
?
Talking about power sensor, I got lucky enough to get a (working) 436b and a WORKING \o/ sensor. But unfortunately, got it without the cable. Is there a cheaper source for the cable than ebay?
---8<---Corte aqui---8<---
---8<---Corte aqui---8<---
?
?
Em qui., 19 de mai. de 2022 às 13:41, visitslovenija <musto102@...> escreveu:
An interesting and long running thread.
Like many, I have a bit of a soft spot for the old HP435 meters and would like to have one working on the bench.
Since broken detectors do show up now and again at a decent price, has anyone successfully reworked one with busted thermocouples and modified for a diode front end? I know it’s small and very fiddly, but is that not easier than trying to build one from scratch?
|
Re: A homemade diode power sensor for HP meters
The 8484A used a single diode and had no temperature compensation , I think that HP tried to minimize the impact of temperature variations by using a heavier all metal case .... the 8481D uses a thermistor
toggle quoted message
Show quoted text
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io <HP-Agilent-Keysight-equipment@groups.io> On Behalf Of Paul Bicknell via groups.io Sent: Thursday, May 19, 2022 4:22 PM To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters OK Dave Re the 8484 I actually have 2 of the black 2 grey and one D version a lot better than my original 430 power meter and Sensor with an N connecter on it So just a thought how dues the 8484A do its temperature compensation as I do not think I have ever looked at the CCT diagram for the 8484 A Paul -----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Lothar baier Sent: 19 May 2022 22:08 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters The 8484A was the first diode sensor that HP produced , it was a horrible piece of equipment prone to drift especially if you got one with the black case . Eventually HP redesigned the sensor to solve the issues and the new sensor was the 8481D On May 19, 2022, at 15:37, Paul Bicknell via groups.io <admin@...> wrote:
?Thank you Dave for those kind words of love Paul
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Dave McGuire Sent: 19 May 2022 21:26 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
On 5/19/22 16:19, Paul Bicknell wrote: Do not worry about Typos as I must be the worst speller on this sight Of that there can be NO doubt. But we all love you anyway. ;)
-Dave
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: A homemade diode power sensor for HP meters
Paul Was in no way trying to solve everyones needs. It was just a terrible thing to waste the nice analog devices detector and the poor old HP435 on the shelf.? With the arduino in there I also get a usb 232 output the prints information. That was just because I could. It wasn't expensive but did require soldering various disassembly and such. Your diode approach sounds like a better answer to get the hp meter closer to teh?original design. Regards Paul.
|
Re: Keysight VNA Cal Kit definitions webpage is missing
All - Did I miss something I'm looking for a copy of the CAL KIT EDITOR - looked at the links that went buy but did not find a copy - does anyone have a copy tot the CAL KIT COEFFICIENTS editor ??
If so, please let me know where.
Cheers!
Bruce
Quoting Jeff Keyzer <jeff@...>:
toggle quoted message
Show quoted text
Caesar,
Thank you so much for making all of this information available for users of the legacy VNAs. I have used your site at various times over the years and it has always been very helpful. Sorry to hear about the hosting issues. I wish I could have grabbed a copy of everything before it started to disappear, but I'll try to save a copy of what's left before it's gone forever.
Thanks again, your efforts are appreciated. Jeff
|
Re: Keysight VNA Cal Kit definitions webpage is missing
Caesar,
I grabbed a copy of this the other day, thank you. I was hoping to find copies of the various cal kit definition zip files that were formerly hosted on the site, as I've used these in the past, but I have since learned that I can recreate these using the tools that are available. I've used the 8753 series intermittently for over twenty years and I am still learning things.
Thanks again, Jeff
|
Re: Keysight VNA Cal Kit definitions webpage is missing
Caesar,
Thank you so much for making all of this information available for users of the legacy VNAs. I have used your site at various times over the years and it has always been very helpful. Sorry to hear about the hosting issues. I wish I could have grabbed a copy of everything before it started to disappear, but I'll try to save a copy of what's left before it's gone forever.
Thanks again, your efforts are appreciated. Jeff
|
Re: A homemade diode power sensor for HP meters
OK Dave
Re the 8484 I actually have 2 of the black 2 grey and one D version a lot better than my original 430 power meter and Sensor with an N connecter on it So just a thought how dues the 8484A do its temperature compensation as I do not think I have ever looked at the CCT diagram for the 8484 A
Paul
toggle quoted message
Show quoted text
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Lothar baier Sent: 19 May 2022 22:08 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters The 8484A was the first diode sensor that HP produced , it was a horrible piece of equipment prone to drift especially if you got one with the black case . Eventually HP redesigned the sensor to solve the issues and the new sensor was the 8481D On May 19, 2022, at 15:37, Paul Bicknell via groups.io <admin@...> wrote:
?Thank you Dave for those kind words of love Paul
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Dave McGuire Sent: 19 May 2022 21:26 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
On 5/19/22 16:19, Paul Bicknell wrote: Do not worry about Typos as I must be the worst speller on this sight Of that there can be NO doubt. But we all love you anyway. ;)
-Dave
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: A homemade diode power sensor for HP meters
The 8484A was the first diode sensor that HP produced , it was a horrible piece of equipment prone to drift especially if you got one with the black case . Eventually HP redesigned the sensor to solve the issues and the new sensor was the 8481D
toggle quoted message
Show quoted text
On May 19, 2022, at 15:37, Paul Bicknell via groups.io <admin@...> wrote:
?Thank you Dave for those kind words of love Paul
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Dave McGuire Sent: 19 May 2022 21:26 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
On 5/19/22 16:19, Paul Bicknell wrote: Do not worry about Typos as I must be the worst speller on this sight Of that there can be NO doubt. But we all love you anyway. ;)
-Dave
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: A homemade diode power sensor for HP meters
Thank you Dave for those kind words of love Paul
toggle quoted message
Show quoted text
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Dave McGuire Sent: 19 May 2022 21:26 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters On 5/19/22 16:19, Paul Bicknell wrote: Do not worry about Typos as I must be the worst speller on this sight Of that there can be NO doubt. But we all love you anyway. ;) -Dave -- Dave McGuire, AK4HZ New Kensington, PA
|
Re: A homemade diode power sensor for HP meters
On 5/19/22 16:19, Paul Bicknell wrote: Do not worry about Typos as I must be the worst speller on this sight Of that there can be NO doubt. But we all love you anyway. ;) -Dave -- Dave McGuire, AK4HZ New Kensington, PA
|
Re: A homemade diode power sensor for HP meters
Hi Bruce Thank you for confirming the 8484A is a Diode front end and On re reading your previews mail again it is as clear as MUD
Do not worry about Typos as I must be the worst speller on this sight Paul
toggle quoted message
Show quoted text
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Bruce Sent: 19 May 2022 21:12 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters It is a type - I meant Diode The 8484 is a DIODE sensor - many other HP diode ensors have a D in the model number as explained B4 Cheers! Beruce Quoting Paul Bicknell <admin@...>: OK Bruce please explain !
You say (The 8484 is a dione power sensor. Many other HP diode sensors end with t "D" instead of an "A" in the model number.)
I do not understand ?? also is dione in the above line a typo
So Is the 8484A power sensor is a diode front end ?? Paul
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Bruce Sent: 19 May 2022 20:45 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
The 8484 is a dione power sensor. Many other HP diode sensors end with t "D" instead of an "A" in the model number.
E.g. 8485D as opposed to 8485A
Cheers!
Bruce
Quoting Paul Bicknell <admin@... <mailto:admin@...> >:
From memory the 8484 power sensor is a diode front end
But a working 8481 or 8484 go for about ?120 to ?160 in the UK equivalent to a day’s wages
And it would take best part of a day to modify busted thermocouples for a diode by the time you did any calibration
Regards Paul
From: HP-Agilent-Keysight-equipment@groups.io <mailto:HP-Agilent-Keysight-equipment@groups.io> [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of visitslovenija Sent: 19 May 2022 11:20 To: HP-Agilent-Keysight-equipment@groups.io <mailto:HP-Agilent-Keysight-equipment@groups.io> Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
An interesting and long running thread.
Like many, I have a bit of a soft spot for the old HP435 meters and would like to have one working on the bench.
Since broken detectors do show up now and again at a decent price, has anyone successfully reworked one with busted thermocouples and modified for a diode front end? I know it’s small and very fiddly, but is that not easier than trying to build one from scratch?
|
Re: A homemade diode power sensor for HP meters
It is a type - I meant Diode
The 8484 is a DIODE sensor - many other HP diode ensors have a D in the model number as explained B4
Cheers! Beruce
Quoting Paul Bicknell <admin@...>:
toggle quoted message
Show quoted text
OK Bruce please explain !
You say (The 8484 is a dione power sensor. Many other HP diode sensors end with t "D" instead of an "A" in the model number.)
I do not understand ?? also is dione in the above line a typo
So Is the 8484A power sensor is a diode front end ?? Paul
-----Original Message----- From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Bruce Sent: 19 May 2022 20:45 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
The 8484 is a dione power sensor. Many other HP diode sensors end with t "D" instead of an "A" in the model number.
E.g. 8485D as opposed to 8485A
Cheers!
Bruce
Quoting Paul Bicknell <admin@... <mailto:admin@...> >:
From memory the 8484 power sensor is a diode front end
But a working 8481 or 8484 go for about ?120 to ?160 in the UK equivalent to a day’s wages
And it would take best part of a day to modify busted thermocouples for a diode by the time you did any calibration
Regards Paul
From: HP-Agilent-Keysight-equipment@groups.io <mailto:HP-Agilent-Keysight-equipment@groups.io> [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of visitslovenija Sent: 19 May 2022 11:20 To: HP-Agilent-Keysight-equipment@groups.io <mailto:HP-Agilent-Keysight-equipment@groups.io> Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
An interesting and long running thread.
Like many, I have a bit of a soft spot for the old HP435 meters and would like to have one working on the bench.
Since broken detectors do show up now and again at a decent price, has anyone successfully reworked one with busted thermocouples and modified for a diode front end? I know it’s small and very fiddly, but is that not easier than trying to build one from scratch?
|
Re: A homemade diode power sensor for HP meters
The connecters are available, Binder make 3 types and they did cost ?10 , ? 13 ?and ?16? each ? Regarding Diode sensors do not forget the 430 meter was made for Diode sensors? also I believe the early 431 A had a switch position for Diode ? Personally looking for a modern meter ( not expensive ) for wave guide diode detectors but calibrated in dB Paul ?
toggle quoted message
Show quoted text
From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Alexandre Souza Sent: 19 May 2022 20:49 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters ? Talking about power sensor, I got lucky enough to get a (working) 436b and a WORKING \o/ sensor. But unfortunately, got it without the cable. Is there a cheaper source for the cable than ebay? ---8<---Corte aqui---8<--- ---8<---Corte aqui---8<--- ? ? Em qui., 19 de mai. de 2022 às 13:41, visitslovenija <musto102@...> escreveu: An interesting and long running thread.
Like many, I have a bit of a soft spot for the old HP435 meters and would like to have one working on the bench.
Since broken detectors do show up now and again at a decent price, has anyone successfully reworked one with busted thermocouples and modified for a diode front end? I know it’s small and very fiddly, but is that not easier than trying to build one from scratch?
|
Re: A homemade diode power sensor for HP meters
OK Bruce? please explain !
You say ? (The 8484 is a dione power sensor.? Many other HP diode sensors end with t "D" instead of an "A" in the model number.)
I do not understand ??? also is dione in the above line a typo
So Is the 8484A power sensor is a diode? front end ??
Paul
toggle quoted message
Show quoted text
-----Original Message-----
From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of Bruce
Sent: 19 May 2022 20:45
To: HP-Agilent-Keysight-equipment@groups.io
Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
The 8484 is a dione power sensor.? Many other HP diode sensors end with t "D" instead of an "A" in the model number.
E.g. 8485D as opposed to 8485A
Cheers!
Bruce
Quoting Paul Bicknell <admin@...>:
> From memory the 8484 power sensor is a diode? front end
>
>
>
> But a working 8481? or 8484 go for about ?120 to ?160 in the UK
> equivalent to a day’s wages
>
> And it would take best part of a day to modify busted thermocouples
> for a diode by the time you did any calibration
>
> Regards Paul
>
>
>
> From: HP-Agilent-Keysight-equipment@groups.io
> [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of
> visitslovenija
> Sent: 19 May 2022 11:20
> To: HP-Agilent-Keysight-equipment@groups.io
> Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power
> sensor for HP meters
>
>
>
> An interesting and long running thread.
>
> Like many, I have a bit of a soft spot for the old HP435 meters and
> would like to have one working on the bench.
>
> Since broken detectors do show up now and again at a decent price, has
> anyone successfully reworked one with busted thermocouples and
> modified for a diode front end? I know it’s small and very fiddly, but
> is that not easier than trying to build one from scratch?
>
>
>
>
>
>
|
Re: A homemade diode power sensor for HP meters
Hi Paul sounds like you have re engendered the 435? and you say you have achieved a 65 dB dynamic range? that is excellent and good for most Amateurs to ?6 Ghz ?but a lot want 10 Ghz? ?suggest you write it up for one of the magazines that Pay just as mater of interest? other than lots of Time ( I assume Weeks ) how much did it cost excluding the donor 435 ? Personally I repair the 432 and 435 meter usually it is the meter face and help UK Amateurs out with parts Paul ?
toggle quoted message
Show quoted text
From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of paulswed Sent: 19 May 2022 18:52 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters ? The diode approach would be the way to go. Like many of you I also like the large scale HP power meters. You can pick the bodies up for nothing. Though suddenly there seems to be a magical market for a body without sensors. An approach I recently took that has worked was to leverage the analog devices power sensors into a 435 body. This approach really couldn't use anything?inside of the 435. The Analog device sensor in a small box with very flexible?cable goes to an arduino that maps the linear to the log scale of the 435. Since the sensor works from?+15 to -50db I reused the switch to select new scaling in 10 db steps. Also to create a broad range 0 to -50 db scale. I like the meter a lot and it now sets on top?the real HP-435. The real one for very accurate readings to 18 GHz the new one good to 6 GHz. It really works well as I rarely need .25 db accuracy on anything. The Linear to log was quite a challenge as I learned. Though the meter is very nice this was a complex project so not sure its useful to many. If the diode approach can work as Dave says then that would be the best approach with the least building challenge. If you don't need 18GHz bandwidth maybe you can get away with larger components.
|
Re: A homemade diode power sensor for HP meters
Talking about power sensor, I got lucky enough to get a (working) 436b and a WORKING \o/ sensor. But unfortunately, got it without the cable. Is there a cheaper source for the cable than ebay?
73 de PU2SEX Alexandre
---8<---Corte aqui---8<---
---8<---Corte aqui---8<--- Em qui., 19 de mai. de 2022 às 13:41, visitslovenija < musto102@...> escreveu:
toggle quoted message
Show quoted text
An interesting and long running thread.
Like many, I have a bit of a soft spot for the old HP435 meters and would like to have one working on the bench.
Since broken detectors do show up now and again at a decent price, has anyone successfully reworked one with busted thermocouples and modified for a diode front end? I know it’s small and very fiddly, but is that not easier than trying to build one from scratch?
|
Re: A homemade diode power sensor for HP meters
The 8484 is a dione power sensor. Many other HP diode sensors end with t "D" instead of an "A" in the model number.
E.g. 8485D as opposed to 8485A
Cheers!
Bruce
Quoting Paul Bicknell <admin@...>:
toggle quoted message
Show quoted text
From memory the 8484 power sensor is a diode front end
But a working 8481 or 8484 go for about ?120 to ?160 in the UK equivalent to a day’s wages
And it would take best part of a day to modify busted thermocouples for a diode by the time you did any calibration
Regards Paul
From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of visitslovenija Sent: 19 May 2022 11:20 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters
An interesting and long running thread.
Like many, I have a bit of a soft spot for the old HP435 meters and would like to have one working on the bench.
Since broken detectors do show up now and again at a decent price, has anyone successfully reworked one with busted thermocouples and modified for a diode front end? I know it’s small and very fiddly, but is that not easier than trying to build one from scratch?
|
Re: Converting a 3336A signal gen to 3336C
Updates & errors in my previous post :
- Display keyboard panel is A5 not A4...
- High accuracy attenuator is still there...its center behind the "modify" potentiometer control...whats is removed is the "balanced output module" designated A11 for the 3336A and A10 for the 3336B
To avoid more "hacking" of the 3336A front panel in the conversion to a 3336C, I cut traces to SW37 & its button led and tacked all this on to SW36 pads. 50 & 75 outputs are not "centered" like the 3336C but this way buttons are aligned with corresponding outputs. I left SW38 (now useless) in. A metal plate was affixed to he back of the front panal to fill the 2 square holes closest to the "modify" pot in front panel...round holes were drilled in that plate and a 50 ohms bnc and a 75 ohms "F" female connector were put there. Claude Claude
|
Re: A homemade diode power sensor for HP meters
From memory the 8484 power sensor is a diode? front end ? But a working 8481? or 8484 go for about ?120 to ?160 in the UK equivalent to a day’s wages? And it would take best part of a day to modify busted thermocouples for a diode by the time you did any calibration Regards Paul ?
toggle quoted message
Show quoted text
From: HP-Agilent-Keysight-equipment@groups.io [mailto:HP-Agilent-Keysight-equipment@groups.io] On Behalf Of visitslovenija Sent: 19 May 2022 11:20 To: HP-Agilent-Keysight-equipment@groups.io Subject: Re: [HP-Agilent-Keysight-equipment] A homemade diode power sensor for HP meters ? An interesting and long running thread.
Like many, I have a bit of a soft spot for the old HP435 meters and would like to have one working on the bench.
Since broken detectors do show up now and again at a decent price, has anyone successfully reworked one with busted thermocouples and modified for a diode front end? I know it’s small and very fiddly, but is that not easier than trying to build one from scratch?
|
Re: hp 16702 LA connectivity questio: GPIB, RS-232 > USB, Ethernet
On Thu, May 19, 2022 at 11:20 AM Mark Litwack <mlitwack@...> wrote: I don't see that there was ever a GPIB option for the 167xx series.
There is actually a TI 9914 GPIB controller populated on the 16700-series logic analyzer CPU board, along with a 75ALS160 and 74ALS164, and a 24-pin header populated where a back panel GPIB connector could have been populated instead. That is because the 16700-series logic analyzer CPU board was originally used for the E4406A. There are E4406-xxxxx stickers on some of the programmable components of the 16700 CPU board. In the E4406A application the back panel GPIB connector is populated on the CPU board. Maybe there were plans to make use of the GPIB controller in the 16700-series logic analyzer application that never materialized.
|