¿ªÔÆÌåÓý

Date

Re: HP 3455a DVM repair

Ed Giaimo
 

My guess is that Q12 is shorted Collector to Base and CR15 is what is pulling TP4 down to -18V and caused R14 to overheat. Q12 is listed as an HP 1854-0087 (75MHz ?360mW NPN).


Re: 8510C questions and cabling

 

OK - You win, but the daisy chain cables are expensive. I'b be tempted to get a long 8502 - test set cable so you could "remote" the test set, which would make exchanging test sets easier.

If you don't already know about them, these guys can make just about any cable you need - not cheapo, but theey do good work -->


Have fun NOT watching TV :-)

Cheers!


Quoting "Reginald Beardsley via groups.io" <pulaskite@...>:

Well, to have an S parameter set, a TR set and an ISM band antenna range.

Do I really need all that? No. But what else am I going to do? Watch TV? Not something that appeals to me.

Have Fun!
Reg


Re: Fixing a sick 4332A LCR meter

 

If you need that FET, I can send you one...

Dan in Chandler, AZ


Re: Fixing a sick 4332A LCR meter

 

Stephen,

1855-0082, TRANSISTOR:FIELD EFFECT P-CHANNEL

Sean


On Fri, Apr 3, 2020 at 04:28 PM, Stephen Hanselman wrote:
Seam,
?
Please send the p/n I¡¯ll check stock

Regards,

?

Stephen Hanselman

Datagate Systems, LLC

?

?
?
?


Re: Fixing a sick 4332A LCR meter

 

¿ªÔÆÌåÓý

Seam,

Please send the p/n I¡¯ll check stock

Regards,

?

Stephen Hanselman

Datagate Systems, LLC




On Apr 3, 2020, at 13:53, [email protected] wrote:

?Hi Chuck,

Good point. So far it's stable. I aligned it as best I am able for now, and it's working in all modes with reasonable accuracy. If it becomes intermittent, looks like it is still possible to obtain these FETs, though it will probably cost.

Seam

On Fri, Apr 3, 2020 at 01:33 PM, Chuck Harris wrote:
Just a repair note:

I have seen a lot of TO92 packaged transistors where
one or more lead has become intermittent. I believe
this is caused by fiddle fingered folks that just have
to push the transistor around to match their ideal of
position, or to read the number.

It always seems to happen when the transistor is mounted
with the leads in a tripod position with the base lead
bent forwards, or backwards.

Touch, or otherwise move the transistor, and it quits...
touch it again, and it works.

-Chuck Harris


Re: 8510C questions and cabling

 

Well, to have an S parameter set, a TR set and an ISM band antenna range.

Do I really need all that? No. But what else am I going to do? Watch TV? Not something that appeals to me.

Have Fun!
Reg


Re: 8510C questions and cabling

 

I found an 08513-60004 on ebay.

Has anyone done an opt 001 upgrade on a non-opt 001 8511A, 8512A or 8515A?

It would appear to be just recabling and adding the multiplexer boards. Of course, I'd also need a 3rd 08510-60102 cable and 2 sets of the opt 001 cables.

Thanks,
Reg


Re: 8510C questions and cabling

 

You might consider why you actually need all three test sets. Multiple test sets does increase complexity and causes losses in the signal from the sig gen.

Cheers!


Quoting "Reginald Beardsley via groups.io" <pulaskite@...>:

The "Preflight Checkout" diagram is clear about where things go, but it presents every variant on the same diagram with cryptic notes. so deciphering the connections for a specific set of modules is difficult.

The application note for using multiple test sets on the 8515A page is a broken link, but I was able to verify from the 8512A operation and service manual that option 001 is needed on all but the last test set in the chain. Unfortunately, neither of mine has that marked on the label. From the 8512A manual it is the A2 board, 08513-60004 (new) or 08513-69004 (rebuilt) and 15 cables.

At present it appears I need:

4x 08512-20019 port extension jumpers or suitable substitutes

2x 08510-60102 test set interconnect cables

08513-6x004 with internal cables

8511A

3x BNC cables for TRIGGER, STOP SWEEP and 10 MHz

a ST3P relay good to 26 GHz and 4x 3.5 mm cables good to 26 GHz.

The BNC cables should not be a big deal so far as I can see. I can live with 18 GHz max if I have to, but it would be nice to get it all up to 26 GHz.

Please send me PM if you have any of this for sale.

Thanks,
Reg


Re: 8510C questions and cabling

 

The "Preflight Checkout" diagram is clear about where things go, but it presents every variant on the same diagram with cryptic notes. so deciphering the connections for a specific set of modules is difficult.

The application note for using multiple test sets on the 8515A page is a broken link, but I was able to verify from the 8512A operation and service manual that option 001 is needed on all but the last test set in the chain. Unfortunately, neither of mine has that marked on the label. From the 8512A manual it is the A2 board, 08513-60004 (new) or 08513-69004 (rebuilt) and 15 cables.

At present it appears I need:

4x 08512-20019 port extension jumpers or suitable substitutes

2x 08510-60102 test set interconnect cables

08513-6x004 with internal cables

8511A

3x BNC cables for TRIGGER, STOP SWEEP and 10 MHz

a ST3P relay good to 26 GHz and 4x 3.5 mm cables good to 26 GHz.

The BNC cables should not be a big deal so far as I can see. I can live with 18 GHz max if I have to, but it would be nice to get it all up to 26 GHz.

Please send me PM if you have any of this for sale.

Thanks,
Reg


Re: 58503A GPSDO PSU details/availability

 

Excellent info thanks Jeremy!

Yes please send wiring details.

I'm hopefully a few weeks away from completing my shack, and freq references are 1st on the list for loading into the rack. :^)

On 4/04/20 5:30 am, Jeremy Owen wrote:
Hi Andy,
I asssume you sorted this out after all this time, but if not, I can send you the wiring details. And it's just a very ordinary PSU.? 5V 1A and +/-15V 0.5A.
Look on EEVBLOG and you will find various comments about psu replacement.
-Jeremy
_._,_._,_


Re: Fixing a sick 4332A LCR meter

 

Hi Chuck,

Good point. So far it's stable. I aligned it as best I am able for now, and it's working in all modes with reasonable accuracy. If it becomes intermittent, looks like it is still possible to obtain these FETs, though it will probably cost.

Seam


On Fri, Apr 3, 2020 at 01:33 PM, Chuck Harris wrote:
Just a repair note:

I have seen a lot of TO92 packaged transistors where
one or more lead has become intermittent. I believe
this is caused by fiddle fingered folks that just have
to push the transistor around to match their ideal of
position, or to read the number.

It always seems to happen when the transistor is mounted
with the leads in a tripod position with the base lead
bent forwards, or backwards.

Touch, or otherwise move the transistor, and it quits...
touch it again, and it works.

-Chuck Harris


Re: Fixing a sick 4332A LCR meter

 

Just a repair note:

I have seen a lot of TO92 packaged transistors where
one or more lead has become intermittent. I believe
this is caused by fiddle fingered folks that just have
to push the transistor around to match their ideal of
position, or to read the number.

It always seems to happen when the transistor is mounted
with the leads in a tripod position with the base lead
bent forwards, or backwards.

Touch, or otherwise move the transistor, and it quits...
touch it again, and it works.

-Chuck Harris

[email protected] wrote:

Joel,

Thanks for the detailed steps on that.

I believe I found the smoking gun, a cold solder joint. I removed A1Q1 to test it out of circuit, which really didn't lead to anything. But it started working when I re-soldered it. I'm going to carry out more testing now and see if I can do the alignment (as best I can anyway, as I don't yet have the standard cap and inductor that are called for.) I think it makes sense now...if this FET wasn't switching correctly I think it would impact the operation of the oscillator.

Sean

On Fri, Apr 3, 2020 at 12:39 PM, Joel Setton wrote:


Sean,

I think we're making progress !

In Capacitance mode, the oscillator should run at 100 kHz in the left-hand
side of the range switch, and at 1kHz on the right-hand ranges. The
frequency switch is driven by relay K1.
Your voltage measurements on both sides of R25 show that when the
oscillator is stopped, the AGC loop (IC2, IC3, Q1) is rightly "attempting"
to restart it by increasing the gain of IC1, i.e. driving A1Q1 to a low
resistance. But A1Q1 is in variable-resistance mode only if its gate
voltage is positive, hence the 2.4V ideal value.

One thing that's puzzling me is that the oscillator should run at 1kHz or
100 kHz, but never at 12 kHz.Your measurements show that it runs at 12 kHz
when it should be doing 1kHz, on the "right-hand" C ranges. Let's see what
failure could increase frequency from an ideal 1kHz to 12 kHz. The 1kHz
frequency is set by R2, R3, C3 and C5. Testing the resistors should be
easy with an ohmmeter, and I would think they are OK, because if one of
them had failed the oscillator would still work OK on 100 kHz, using R1
and R4 (but it doesn't).
Testing the capacitors is more difficult. I would think that C3 and C5 are
suspects. You may want to try this: set the meter in Capacitance mode to
the "right" ranges, when the oscillator runs at 12 kHz. Then shunt C3 with
a 1nF capacitor and look at the output frequency (12kHz or 1 kHz ?). Then
remove the 1nF shunt capacitor across C3, and put a 10nF capacitor across
C5. Again, look at the oscillator frequency (12 kHz or 1kHz ?). That will
tell us something.

If one of capacitors C3 or C5 has failed in "open" mode, it would explain
why the oscillator runs at 12 kHz instead of 1 kHz. It would also explain
why it doesn't run at all on the "left-hand" ranges because while it can
do 100kHz, it can't run at 1 MHz. IC1 and IC2 are just too slow for that.

And, as before, DO keep me posted !

Joel



Re: Fixing a sick 4332A LCR meter

 

Joel,

Thanks for the detailed steps on that.

I believe I found the smoking gun, a cold solder joint. I removed A1Q1 to test it out of circuit, which really didn't lead to anything. But it started working when I re-soldered it. I'm going to carry out more testing now and see if I can do the alignment (as best I can anyway, as I don't yet have the standard cap and inductor that are called for.) I think it makes sense now...if this FET wasn't switching correctly I think it would impact the operation of the oscillator.

Sean


On Fri, Apr 3, 2020 at 12:39 PM, Joel Setton wrote:
Sean,

I think we're making progress !

In Capacitance mode, the oscillator should run at 100 kHz in the left-hand side of the range switch, and at 1kHz on the right-hand ranges. The frequency switch is driven by relay K1.
Your voltage measurements on both sides of R25 show that when the oscillator is stopped, the AGC loop (IC2, IC3, Q1) is rightly "attempting" to restart it by increasing the gain of IC1, i.e. driving A1Q1 to a low resistance. But A1Q1 is in variable-resistance mode only if its gate voltage is positive, hence the 2.4V ideal value.

One thing that's puzzling me is that the oscillator should run at 1kHz or 100 kHz, but never at 12 kHz.Your measurements show that it runs at 12 kHz when it should be doing 1kHz, on the "right-hand" C ranges. Let's see what failure could increase frequency from an ideal 1kHz to 12 kHz. The 1kHz frequency is set by R2, R3, C3 and C5. Testing the resistors should be easy with an ohmmeter, and I would think they are OK, because if one of them had failed the oscillator would still work OK on 100 kHz, using R1 and R4 (but it doesn't).
Testing the capacitors is more difficult. I would think that C3 and C5 are suspects. You may want to try this: set the meter in Capacitance mode to the "right" ranges, when the oscillator runs at 12 kHz. Then shunt C3 with a 1nF capacitor and look at the output frequency (12kHz or 1 kHz ?). Then remove the 1nF shunt capacitor across C3, and put a 10nF capacitor across C5. Again, look at the oscillator frequency (12 kHz or 1kHz ?). That will tell us something.

If one of capacitors C3 or C5 has failed in "open" mode, it would explain why the oscillator runs at 12 kHz instead of 1 kHz. It would also explain why it doesn't run at all on the "left-hand" ranges because while it can do 100kHz, it can't run at 1 MHz. IC1 and IC2 are just too slow for that.

And, as before, DO keep me posted !

Joel


Re: Fixing a sick 4332A LCR meter

 

Sean,

I think we're making progress !

In Capacitance mode, the oscillator should run at 100 kHz in the left-hand side of the range switch, and at 1kHz on the right-hand ranges. The frequency switch is driven by relay K1.
Your voltage measurements on both sides of R25 show that when the oscillator is stopped, the AGC loop (IC2, IC3, Q1) is rightly "attempting" to restart it by increasing the gain of IC1, i.e. driving A1Q1 to a low resistance. But A1Q1 is in variable-resistance mode only if its gate voltage is positive, hence the 2.4V ideal value.

One thing that's puzzling me is that the oscillator should run at 1kHz or 100 kHz, but never at 12 kHz.Your measurements show that it runs at 12 kHz when it should be doing 1kHz, on the "right-hand" C ranges. Let's see what failure could increase frequency from an ideal 1kHz to 12 kHz. The 1kHz frequency is set by R2, R3, C3 and C5. Testing the resistors should be easy with an ohmmeter, and I would think they are OK, because if one of them had failed the oscillator would still work OK on 100 kHz, using R1 and R4 (but it doesn't).
Testing the capacitors is more difficult. I would think that C3 and C5 are suspects. You may want to try this: set the meter in Capacitance mode to the "right" ranges, when the oscillator runs at 12 kHz. Then shunt C3 with a 1nF capacitor and look at the output frequency (12kHz or 1 kHz ?). Then remove the 1nF shunt capacitor across C3, and put a 10nF capacitor across C5. Again, look at the oscillator frequency (12 kHz or 1kHz ?). That will tell us something.

If one of capacitors C3 or C5 has failed in "open" mode, it would explain why the oscillator runs at 12 kHz instead of 1 kHz. It would also explain why it doesn't run at all on the "left-hand" ranges because while it can do 100kHz, it can't run at 1 MHz. IC1 and IC2 are just too slow for that.

And, as before, DO keep me posted !

Joel


Re: 8510C questions and cabling

 

You can daisy chain the test sets IF the test sets have the proper internal connections. Look to see if both "D" connectors have coaxial fittings, if so you can daisy chain. Test sets with this option are relatively rare in my experience.

There is an HP note about using multiple test sets with an 8510 - I'll look in my library to see if I can find it.
Quoting Pete Manfre <pmanfre@...>:

Yes you can't daisy chain the but you need option 001 in one of the test
sets...I do believe.

Pete wa2odo

On Fri, Apr 3, 2020, 1:49 PM Reginald Beardsley via groups.io <pulaskite=
[email protected]> wrote:

None of the cabling is obvious from the Preflight Checkout diagram or the
entire document. At this point my concern is identifying specialty HP
cables I need to get.

I'd like to be able to switch the 83631B among an 8511A, 8512A and 8515A
so I don't have to recable the system to change modes of operation. That
seems pretty straight forward, add a ST3P RF relay. But I can't find
anything to tell me if I can daisy chain the 8512A and 8515A using a
08510-60102 cable between the two. The presence of the J10 and J11
connectors on both test sets suggest you can, but I've not yet found any
supporting documentation.

And then there are mysteries such as, "For 8360, a BNC open, 1250-0053
should be connected to the 85102 SWEEP IN 0-10V". How is that different
from the bare BNC connector or a BNC dust cap?

I'm installing everything in custom racks built above a pair of ESD
surface benches. It's a small room, so pulling the bench out from the
wall to recable is not attractive if it's possible to avoid it.

Reg




Re: 8510C questions and cabling

 

The statement about the BNC open is an awkwardly worded
statement, likely by an English challenged individual.

All they are trying to say is don't connect anything to
that BNC connector... especially don't short it, or use
a 50 ohm terminator!

The 83631 trades a bnc cable for one of those combo RF and
multiple pin "D" style connectors... only they are using
only the signal pins, not the coax pins.

Was the cabling notes I posted sufficiently clear?

It described the cabling from the point of view of each
instrument looking outwards. In other words, you should
be able to look at the back of your test set, and see
from my description where each cable goes... Or look at
the back of your display unit, and see from my description
where each cable goes...

-Chuck Harris

Reginald Beardsley via groups.io wrote:

None of the cabling is obvious from the Preflight Checkout diagram or the entire document. At this point my concern is identifying specialty HP cables I need to get.

I'd like to be able to switch the 83631B among an 8511A, 8512A and 8515A so I don't have to recable the system to change modes of operation. That seems pretty straight forward, add a ST3P RF relay. But I can't find anything to tell me if I can daisy chain the 8512A and 8515A using a 08510-60102 cable between the two. The presence of the J10 and J11 connectors on both test sets suggest you can, but I've not yet found any supporting documentation.

And then there are mysteries such as, "For 8360, a BNC open, 1250-0053 should be connected to the 85102 SWEEP IN 0-10V". How is that different from the bare BNC connector or a BNC dust cap?

I'm installing everything in custom racks built above a pair of ESD surface benches. It's a small room, so pulling the bench out from the wall to recable is not attractive if it's possible to avoid it.

Reg




Re: 8510C questions and cabling

Pete Manfre
 

Yes you can't daisy chain the but you need option 001 in one of the test sets...I do believe.?

Pete wa2odo?

On Fri, Apr 3, 2020, 1:49 PM Reginald Beardsley via <pulaskite=[email protected]> wrote:
None of the cabling is obvious from the Preflight Checkout diagram or the entire document.? At this point my concern is identifying specialty HP cables I need to get.

I'd like to be able to switch the 83631B among an 8511A, 8512A and 8515A so I don't have to recable the system to change modes of operation.? That seems pretty straight forward, add a ST3P RF relay.? But I can't find anything to tell me if I can daisy chain the 8512A and 8515A using a 08510-60102 cable between the two.? The presence of the J10 and J11 connectors on both test sets suggest you can, but I've not yet found any supporting documentation.

And then there are mysteries such as, "For 8360, a BNC open, 1250-0053 should be connected to the 85102 SWEEP IN 0-10V".? How is that different from the bare BNC connector or a BNC dust cap?

I'm installing everything in custom racks built above a pair of ESD surface benches.? ?It's a small room, so pulling the bench out from the wall to recable is not attractive if it's possible to avoid it.

Reg




Re: 8510C questions and cabling

 

None of the cabling is obvious from the Preflight Checkout diagram or the entire document. At this point my concern is identifying specialty HP cables I need to get.

I'd like to be able to switch the 83631B among an 8511A, 8512A and 8515A so I don't have to recable the system to change modes of operation. That seems pretty straight forward, add a ST3P RF relay. But I can't find anything to tell me if I can daisy chain the 8512A and 8515A using a 08510-60102 cable between the two. The presence of the J10 and J11 connectors on both test sets suggest you can, but I've not yet found any supporting documentation.

And then there are mysteries such as, "For 8360, a BNC open, 1250-0053 should be connected to the 85102 SWEEP IN 0-10V". How is that different from the bare BNC connector or a BNC dust cap?

I'm installing everything in custom racks built above a pair of ESD surface benches. It's a small room, so pulling the bench out from the wall to recable is not attractive if it's possible to avoid it.

Reg


Re: HP 3455a DVM repair

 

On 4/3/20 12:33 PM, Bob Albert via groups.io wrote:
After replacing the resistor all seems okay in that area.? No other
failures.? Just a bad resistor that went up in value and then overheated.
Went *up* and overheated? That doesn't sound right. I'd bet that it
overheated and went up in value as a result.

The latest is that a new failure has surfaced.? Now the unit blows
fuses.? One of the regulators gets hot so I have to stop with the
original issue and get into the power supply.? Suddenly the voltages are
out of specification.
Are there tantalum capacitors in a 3455A? (I had one for years, long
ago, but never had to work on it.)

I am getting a bit annoyed with this repair job and so may put it back
into the closet for the nonce.
You're making progress, though!

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Fixing a sick 4332A LCR meter

 

On Fri, Apr 3, 2020 at 09:37 AM, Joel Setton wrote:
Ok, then I'd suggest we focus on getting the oscillator running again !
Since it used to be OK, you may want to start by looking for any broken wires, etc ... Maybe it's just a broken wire.

The problem could logically be a power supply fail, or in the oscillator itself which is on A1.
First check the power supplies; then if they're OK, look at what's happening in the oscillator.
The oscillator only depends on power supply lines, and needs no other signals to operate properly.
The troubleshooting tree (figure 5-8) provides some help in troubleshooting the oscillator, if you need it. It's a conventional Wien oscillator, so you may want to start by checking the DC levels on A1TP1 (approx -0.6V), then on either side of A1R25 which will tell you what the AGC loop is doing.

And please... keep me informed !
Have fun,
Joel
Joel,

I have been following the troubleshooting tree. With the settings suggested (C, 100 pF range, 100 pF cap across the unknown) So far, it's gotten me basically across the bottom of the tree, though I haven't tried disconnecting the feedback and inserting my own bias yet.

A couple things I have observed since the last message:

1) sometimes the K1 reed relay doesn't seem to be switching, other times it does.
2) The oscillator runs in whatever state it is in, on R, in every range.
3) The oscillator stops in C, when the range is in the left hand side of the knob range; it starts (again with the 12 KHz 70 mVrms) in the right hand side of the switch range. The DC offset on A1TP1 is +6 Vdc when the oscillator is off; when it is running it is -0.9 Vdc. Pin 6 of A1IC3 (connected to one side of A1R25) is -12.3 Vdc (supposed to be +2.4 according to schematic???) and the other side of A1R25 is -0.68 Vdc.
4) The behavior in (3) is apparently identical in the L function.

Sean