开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育
Date

Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Hi Razvan,

I have option 124 on my hp53132A and it's a clone bought on eBay.
The scale used is 4, because by default, the internal divider is positioned
for a 030 or 3GHz option.

But I found on the EEVblog site, it seems to me, the following file
concerning the activation of the correct divider depending on the option.
I'll try this approach when I have some free time.


Yves

-----Message d'origine-----
De?: [email protected]
<[email protected]> De la part de Razvan Popescu via
groups.io
贰苍惫辞测é?: 5 mai 2024 16:58
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Another 3GHz 53132-68003 Chanel
3 option design for the 53131A and 53132A

Hello Ed,

I saw how the 3GHz board works regarding the divider but I didn't see any
original board more than 3GHz.

I didn't buy any cloned boards because I saw you need to use the math
function scale set to 4.0. I think they do divide by 128 also for the 12GHz
board but they should do probably divide by 512.

How is the original 12Ghz prescaler board working in regards to the divider?

I got the PCBs for the 3GHz version but I didn't order the parts to build
them. I will install the 3GHz prescaler first until I find a prescaler that
doesn't need the scale set to 4.

Razvan


On 04/05/2024 13:22, Ed (scskits) via groups.io wrote:
Razvan:

The 3GHz board is a divide by 128, with a max output of 23.4375MHz at
3GHz.
That has to get to the main board through the ribbon cable, not the
best way to connect.
At 12GHz the output would be almost 100MHz so I think that the higher
frequency option boards likely have an additional divider to keep the
output at a reasonable frequency.

There are two pins (CH3CODE0 and CH3CODE1) on the connector that are
used to tell the main board that one of the options is installed.

I would think that a combination of those pins would select one of the
three options or no option.

I will try to see what effect they have later this weekend.

ed


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Hello Ed,

I saw how the 3GHz board works regarding the divider but I didn't see
any original board more than 3GHz.

I didn't buy any cloned boards because I saw you need to use the math
function scale set to 4.0. I think they do divide by 128 also for the
12GHz board but they should do probably divide by 512.

How is the original 12Ghz prescaler board working in regards to the divider?

I got the PCBs for the 3GHz version but I didn't order the parts to
build them. I will install the 3GHz prescaler first until I find a
prescaler that doesn't need the scale set to 4.

Razvan

On 04/05/2024 13:22, Ed (scskits) via groups.io wrote:
Razvan:

The 3GHz board is a divide by 128, with a max output of 23.4375MHz at 3GHz.
That has to get to the main board through the ribbon cable, not the best
way to connect.
At 12GHz the output would be almost 100MHz so I think that the higher
frequency option boards likely have an additional divider to keep the
output at a reasonable frequency.

There are two pins (CH3CODE0 and CH3CODE1) on the connector that are
used to tell the main board that one of the options is installed.

I would think that a combination of those pins would select one of the
three options or no option.

I will try to see what effect they have later this weekend.

ed


Re: HP 3852A firmware update issue

 

Rick
I did follow the? instructions on and that seems to be the issue.?? I tried another version from the file section that was 4.21 copied from 27c512 ROMS. The chips need to be 27C512? not 27C256 as mentioned on the www.dalton.ax page or on the notes in the file section .? I uploaded the 4.21 version to 8 chips of 27C512 and all is good - no more warning messages.

The working version 4.21? was posted by ChrisBeee here - make sure to program to 27c512 chips. :
/g/HP-Agilent-Keysight-equipment/files/All%20HP,%20Agilent%20and%20Keysight%20instruments%20in%20folders%20by%20part%20numbers/3000%20to%203999/3852A%20Data%20Acquisition%20Control%20Unit/3852A-421-ROMs.zip
Follow the instructions from HP :
/g/HP-Agilent-Keysight-equipment/files/All%20HP,%20Agilent%20and%20Keysight%20instruments%20in%20folders%20by%20part%20numbers/3000%20to%203999/3852A%20Data%20Acquisition%20Control%20Unit/3852A-4.22-ROM-upgrade.pdf

Too bad the 4.22 is not working.


Maciej


Re: Attenuator "O" Rings

 

I Measured:

The Diameter of the pin itself to be 40 Thousands of an inch which is very close to ONE mm.
This corresponds to the ID of the O Ring
The OD is twice that so the Cross Section is .5 mm..? OD= 2XCS + ID

So I think VK2BEA is quite correct.

I do not have Metric measurement tools so I have to translate from the measurements on my English micrometers and calipers


Re: Attenuator "O" Rings

 

I used these o-rings from 'The O-Ring Store' on the attenuator from the 8595E (a 33321 attenuator I think).

https://www.theoringstore.com/store/index.php?main_page=product_info&cPath=368_12_2185&products_id=43455


Re: Attenuator "O" Rings

 

Do these look correct???


https://www.ebay.com/itm/293851527865?itmmeta=01HX2WM6F270SAZY0X9KSYJH67&hash=item446aea7ab9:g:4NcAAOSwC7Vh1id6&itmprp=enc%3AAQAJAAAA4NyxuU3mlUEZV%2Bn9csEzewYdmbqpmPKU2zfXTNMFqmn8eaEDgCi3paQRuFWeIRfM%2FDoLlVjIxoJ%2FR0dl8MmaBF9jq3BOVXcW20QC4dgpmEEGmh3sQP%2FKY6GaYJNwisDDlQTrSR2GMhOtwoz5QboqhdRA9j45ejvMbJ86V21tFz3d2gJNUfYPaBpMGrUdCRVQLnB%2FqwIN7EQZAhaSXHi5ZNV--jLUlx1cBNT%2B0fVai4RYtWrI3BBxl0dkpZDeX0DQ1fCJpGK0SDf4xeXLO9VkA2xw6bRFIwty%2B4WpTsGf%2BvMW%7Ctkp%3ABk9SR9bn0NzoYw


Re: Attenuator "O" Rings

 

On Sat, May 4, 2024 at 08:51 AM, Radu Bogdan Dicher wrote:
I've used Amazon as source of o-rings when I repaired my 8657B. But to your pertinent point, I think I'd go McMaster-Carr or similar today:?.
McMaster-Carr is also my normal go-to for quality O-rings. I checked there, but they don't go as small as the 1 mm ID/2 mm OD that others called out. Buna-N is certainly the best choice of material.

I don't know about this application. Would the durometer (hardness) matter here? Buna-N O-rings can be bought in a range of durometers, where 70 is the norm, but harder and softer versions are available.

Try Apple Rubber. They have LOTs of choices in small sizes.?


Re: HP 400H AC VTVM Difficulties

 

The 400H is just the higher precision version of the HP 400 family, which had a VERY long history with HP, probably over 30 years, so there are lots of versions, usually differentiated by changes in the serial prefix. (The earliest models did not have a prefix.)

The very first thing to do with any member of this family is to replace all 4 of the large cylindrical capacitors. If yours are still the originals, they will be leaky by now. One of them is in the PS and minor leakage won't be a problem, but you might as well replace all of them while you're at it.

The other 3 are interstage plate to grid coupling capacitors. Leakage here will raise the grid voltage of the following stage, pushing that stage toward saturation, poor performance, lots of heat dissipation, and reduced tube life. Good quality Mylar caps make better than original replacements. Each of these three also has a mica cap in parallel with it, and sometimes I've found one of those to also be leaky. Check each grid voltage to see if it agrees with the schematic.

Once you get these fixed, you can move on to the problem you asked about. I don't have any specific recommendations for that problem, but I'd suspect a problem with poor contact in the range switch. Treatment with Deoxit followed by Shield should be good for this.

If you continue to have problems, post your serial numbers/prefixes so we know what version you're working with. I probably have 5 different manuals for the HP400 series, and I'm sure I'm still not completely covered.


Re: BOONTON 8900B PEAK POWER CALIBRATOR

 

开云体育

To avoid trampling over another group bandwidth i suggest you as this on the Marconi group.io pages
for this question.

And definatly off brand topic, has anyone any details/specs on a Marconi TM9701 VHF/UHF Detector, test accessory. Its an accessory with both a positive and negative output via seperate diodes. Id like to test it but manufacturers specs would be handy.

I suspect it may be possible to help you out.

Regards?


On 3 May 2024, at 22:17, Pete Harrison via groups.io <rockfisher@...> wrote:

?

Hi all, i appreciate this isnt a HPAK instrument but being Boonton it kinda is :)

Has anyone got a full service manual for these please. Ive managed to find one online but it doesnt have a circuit diagram.

I also have a pdf of Boontons no35 notebook? by G.Raymond. Polen who designed the instrument, but id really like the full manual if possible and it will save me trying to draw out the circuit by hand.

?

And definatly off brand topic, has anyone any details/specs on a Marconi TM9701 VHF/UHF Detector, test accessory. Its an accessory with both a positive and negative output via seperate diodes. Id like to test it but manufacturers specs would be handy.

Many thanks in advance Pete Harrison.


HP 3852A firmware update issue

 

I have a 3852A with firmware 3.0 and wanted to use a card that requires firmware 4.2 . I found the instructions and firmware 4.22. I identified my board as 66503 revA? per PCB . Per instructions I only needed to hook pins 1 of U57 and U58 to pin 4 of RP4. I did that but I get a error message that the ROM update is not complete.?
Anyone with revA of 66503 board had this happen?
The suggested remedy on the display is not possible on Rev A boards as there is no jumper wire .?



Re: Attenuator "O" Rings

 

PAI had a minimum order of 100, but when I went up to 500 they started to seem really worth it.? The worst thing about doing a bulk buy was having to count out all those little bags of 50 :^)? I've had no complaints since.
Dan


Re: Attenuator "O" Rings

 

I asked Precision for a quote, I guess we will see Monday.

Wally


Re: BOONTON 8900B PEAK POWER CALIBRATOR

 

Many thanks David, somehow i missed that in my search.

Regards Pete


Re: HP 400H AC VTVM Difficulties

 

开云体育

Since you're okay on the 1mV to 300mV ranges and the trouble starts when you advance to the 1V range, I'm staring at your x1000 range attenuator.? Look for an open there.
What's your serial number?? There were several distinct versions of the circuit as time went on, and the range attenuator was one of the blocks affected.? The early version was known for high noise level.

I've attached repair notes for my 400D which is the same as the 400H except for a lower quality meter movement.? Mine is Serial Prefix 017.? The notes include a partial bibliography of different versions of the manual.

Dave Wise


From: [email protected] <[email protected]> on behalf of SCMenasian <scm@...>
Sent: Saturday, May 4, 2024 5:49 AM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] HP 400H AC VTVM Difficulties
?
Since the problem appears to be a residual reading at the lowest ranges and the grid of V2 helps, you might want to rule out spurious noise getting to V2. Multi MHz oscillation in the V1 stage comes to mind.
I'd check (with a 50 MHz or better 'scope) at the grid of V2. The "phase of the moon" observation supports this hypothesis. Perhaps one or more of the capacitors in the V1 circuit is bad.


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

开云体育

Hi Ed,

?

Both 53132A do not count if there is no signal at their CH3 input.

?

I bought these 2 PCBs on eBay, I think they are clones.

?

For 12.4GHz option, the board is identical to this one:

For the 3GHz option, I cannot find the information.

?

Yves

De?: [email protected] <[email protected]> De la part de Ed (scskits)
贰苍惫辞测é?: 4 mai 2024 07:11
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

?

Yves:

Thanks for the update.

Do the boards count randomly when there is no input?

Were these the OEM options or clones?

ed


Re: Attenuator "O" Rings

 

开云体育

I think precision associates wanted somewhere between 30 and 40 cents each but I don’t recall the minimum order quantity.

About a year ago, they were around 1/3 of the price of apple rubber (NY).


Re: Attenuator "O" Rings

 

Interesting idea, Adrian. I've used Amazon as source of o-rings when I repaired my 8657B. But to your pertinent point, I think I'd go McMaster-Carr or similar today:?. Their products are very good and well classified, spec'ed, and extremely finely organized for searching.?
I'd definitely go this route should I have to do it again.?
Radu.?


On Sat, May 4, 2024 at 2:29?AM Adrian Nicol via <Adrian=[email protected]> wrote:
In metric-land I got a bunch of 1mm ID 2mm OD watch stem seals, eBay or Amazon I seem to recall. That was about 10 years back on an 8566A - still working fine - done several other attenuators since.
However while they seem OK at the moment, nowadays having had cheap plumbing 'rubber' seals go gooey after about 5 years, I have little faith in unbranded stuff and I'll bet it's just a matter of time until they start to return to their original liquid hydrocarbon form.
Probably worth finding some of that size from a known (and trusted) manufacturer?

On 04/05/2024 00:08, Wallace Gasiewicz via wrote:

Does anyone know what size and where I could get "O" rings for my HP 8714 VNA.Attenuator??
One is not there and attenuator is not working properly.

wally KC9INK


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

I have the latest version of my PCB's ready to build. I added a trimpot for the bias on the prescaler chip to do away with the need for selecting a resistor, which is accessible through a convenient spare hole in the side of the chassis.
This way you can easily dial in the bias to optimise the tradeoff between stopping random counts with no input signal and sensitivity of the prescaler.

Here is the sensitivity of my previous version (pdf), once I have my latest version constructed, I'll make a better measurement of its performance.
It exceeds the HP spec by a bit up the top of the frequency range, so I'll have to figure out how to fix that.


And a 3D render of the design:


Re: HP 400H AC VTVM Difficulties

 

Since the problem appears to be a residual reading at the lowest ranges and the grid of V2 helps, you might want to rule out spurious noise getting to V2. Multi MHz oscillation in the V1 stage comes to mind.
I'd check (with a 50 MHz or better 'scope) at the grid of V2. The "phase of the moon" observation supports this hypothesis. Perhaps one or more of the capacitors in the V1 circuit is bad.


Re: HP 400H AC VTVM Difficulties

 

Have a look for the 400D & 400L on here too, they are almost identical to the 400H.

Looking at the manual, it lists a number of paper capacitors, these will almost certainly be leaky, some are grid coupling capacitors.
Also from experience I've had a number of the 1500uF low voltage capacitors fail open circuit, these were single cans rather than the dual can used in this, they were in a pair of 412A and another in a 5245L.

David