开云体育

Date

Re: massive attachments

 

开云体育

This is exactly the reason why many groups do not allow attachments.

DaveD

On Feb 23, 2022, at 14:02, Larry McElhiney via groups.io <lmcelhiney@...> wrote:

?On Wed, Feb 23, 2022 at 12:33 PM, Roy J. Tellason, Sr. wrote:
I'd appreciate it if you guys could simply put such large files where somebody can get at them, and provide a link, for those that want it, rather than forcing those of us who get things this way to download for no immediate benefit.

Hi Roy,

In your subscription to this group, go to Advanced and you will see the maximum attachment size option, below:


Max Attachment Size

?? ?? ?? ?? ? Unlimited ?? ?? ?? ?? ? 0 (no attachments) ?? ?? ?? ?? ? 100KB ?? ?? ?? ?? ? 500KB ?? ?? ?? ?? ? 5MB ? ? ?? ?? ?

Maximum size of attachments sent to you in emails. For attachments larger than this size, links to retrieve them will be provided instead.


You can control the attachments which are sent to you.

Also, you can change Digest or Summary rather than All Messages, which allows you to choose individual messages.


Take care,

Larry


How can I measure the bandwidth of an audio transformer with the HP8903B ?

 

Good day everyone,
I would like to measure the bandwidth of an audio output transformer for a push-pull tube amplifier, using the HP8903B audio analyzer, but I have some doubts about how to connect the transformer to the input and output of the HP8903B.
The primary of the transformer (anode - anode):? do I connect them to the HP8903B output (set to 50 ohm or 600 ohm)?
Do I connect an 8 ohm resistor and the input of the HP8903B to the 8 ohm secondary of the transformer?
What voltage should I set for the HP8903B output?

Thanks for the attention
--.Cheers
??? Attilio
? ?


Re: HP 8594E upgrade CRT to LCD

 

开云体育

Yes, and no.

The EL panel is quite thin and should be relatively easy to physically replace on some of the displays I've seen.? Now you have to find the replacement.

It's possible that the plastic diffuser panel from an LED backlit display might fit, but that may be a difficult thing to do, both to find the panel and to replace it.

You may be able to find a replacement by googling for EL replacement for that display.

I had a tektronix TM5000 function generator with a bad EL panel, and ended up finding an equivalent LED backlit panel.? I actually think that the reverse background looks better.

It might be possible to find a later LCD display with the same mechanical size that uses LEDS or a CCFL.? Bear in mind that the inverter for the CCFL is 900 to 1.5 KV, and the EL is 100 or so.? Be very careful with the CCFL inverter.? I'd personally avoid them and their displays wherever possible.? It is possible to replace a CCFL lamp with an LED array, though.? Some of the replacements are commercially made.

Harvey


On 2/23/2022 4:28 PM, Dave Brown wrote:

Is it possible to give the EL display backlights any new lease of life?? HP used them in a number of instruments and most ‘fixes’ I’ve seen so far mean a replacement.

DaveB, ?NZ

?

From: [email protected] [mailto:[email protected]] On Bethalf Of Harvey White
Sent: Thursday, February 24, 2022 09:46
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 8594E upgrade CRT to LCD

?

From what I've seen, the displays are from a different era, and that could make a significant difference.? Modern TFT displays seem to be a lot more durable.

On the old displays, the EL backlight gives out quickly.

Also not sure about the temperature tolerance, either, although LCD displays do not really like getting too hot.

Harvey

On 2/23/2022 3:09 PM, paulswed wrote:

Stuart not to throw stones. But I have seen a fair number of tek lcd displays that have gone to heck. Normally something like crystal leakage. (I just made that up). I can think of things like the tek TDR 1503 non crt versions yet the crt versions still work.

Just saying we hobbiest users use the gear far past any expectation that it should work.

Regards

Paul

WB8TSL


Re: HP 8594E upgrade CRT to LCD

 

开云体育

Is it possible to give the EL display backlights any new lease of life?? HP used them in a number of instruments and most ‘fixes’ I’ve seen so far mean a replacement.

DaveB, ?NZ

?

From: [email protected] [mailto:[email protected]] On Behalf Of Harvey White
Sent: Thursday, February 24, 2022 09:46
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 8594E upgrade CRT to LCD

?

From what I've seen, the displays are from a different era, and that could make a significant difference.? Modern TFT displays seem to be a lot more durable.

On the old displays, the EL backlight gives out quickly.

Also not sure about the temperature tolerance, either, although LCD displays do not really like getting too hot.

Harvey

On 2/23/2022 3:09 PM, paulswed wrote:

Stuart not to throw stones. But I have seen a fair number of tek lcd displays that have gone to heck. Normally something like crystal leakage. (I just made that up). I can think of things like the tek TDR 1503 non crt versions yet the crt versions still work.

Just saying we hobbiest users use the gear far past any expectation that it should work.

Regards

Paul

WB8TSL


Re: Storage Cap Ripple

 

Thanks for the tip. I may have already replaced this cap. I went through all the usual suspects first time around and replaced all the suspect ones. But I'll double check I've done the one you mentioned; many thanks.
Yes, luck trumps all!


Re: HP 8594E upgrade CRT to LCD

 

开云体育

From what I've seen, the displays are from a different era, and that could make a significant difference.? Modern TFT displays seem to be a lot more durable.

On the old displays, the EL backlight gives out quickly.

Also not sure about the temperature tolerance, either, although LCD displays do not really like getting too hot.

Harvey

On 2/23/2022 3:09 PM, paulswed wrote:

Stuart not to throw stones. But I have seen a fair number of tek lcd displays that have gone to heck. Normally something like crystal leakage. (I just made that up). I can think of things like the tek TDR 1503 non crt versions yet the crt versions still work.
Just saying we hobbiest users use the gear far past any expectation that it should work.
Regards
Paul
WB8TSL


HP6060A questions

 

Just received and works great. But a couple of questions for the group:

(1) Is it possible to add some sort of backlight to the LCD?
(2) My binding posts are in the back. How hard is it to move these to the front?

Thanks in advance.

Mark


Re: Storage Cap Ripple

 

开云体育

Regarding the Off-Lock YTO loop, I suggest you to look at C7 on A20.? This capacitor is known to go dry and create the ugly YTO unlock problem.? A nice gentlemen create a YouTube video about this problem.? I was lucky, as it fixed my 8566B problem.? Any day, I prefer to be lucky than smart ?

?

Best, Francesco K5URG

?

From: [email protected] <[email protected]> On Behalf Of benj3867 via groups.io
Sent: Wednesday, February 23, 2022 5:35 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Storage Cap Ripple

?

@Jinxie, I did not mean to discourage you by saying that this repair is impossible without very accurate equipment that you do not have; this is not necessarily the case, and I myself have fixed many pieces of equipment with surprisingly inadequate gear. What I did want to convey to you is the fact that the 8566B is an extremely sophisticated piece of equipment that requires a precision mindset if one wants to achieve 100% success in troubleshooting it. Assuming that the 22V rail is “good enough” when it measures +1V, and the specified allowed deviation is at most 50 times less (+20mV), did not look like a promising start.

To make my point even more concrete, as Matt said, one can sometimes get rid of the YTO unlock error by performing the “13. Sweep, DAC, and Main Coil Driver Adjustments”. One of the steps in this procedure calls for adjusting the “scan reset offset” voltage to zero with an accuracy of +-500 micro-volts.
In other words, you can not assure a successful repair without accurate gear and mindset, but you may still be lucky enough and be able to fix it with an accurate mindset and less accurate gear.

As for your comment that “no way can I meet those kind of standards...”, I do not believe that this is necessarily the case given that you also say that “I have a *ton* of high-end test gear here..” --- even if it is all vintage with unknown calibration status. For example, my experience with a few used HP 3456A meters is that they are within specs, or very close to it, many years after being last calibrated. Also, even if out-of-spec, their errors are quite consistent and they still have tons of resolution which is what counts most in this case. In other words, your 34566B will function perfectly well if it is adjusted with all voltages consistently off by the same small error: the relative voltages at different points in the circuit are more important than their absolute values (within reason of course).

Finally, if you are lucky, the problem may be simply a bad electrolytic capacitor on A20, A19 or A21. Given that you have an LCR and ESR meters, why don't you start there first?


Re: HP 1727A CRO: Problem with the vertical output amplifier.

 

Thanks for the suggestions, I will look into it further. I should have mentioned that the vertical deflection amplitude is off, as displayed on the screen. I don't know weather that is because of an adjustment or the vertical amp, so a problem here would explain some of it.

I didn't realize the scales were different on the plot, rookie mistake.


Re: eCal & HP-8753E

 

开云体育

I am at 07.74 on Firmware, so I should be ok. ?

On Feb 23, 2022, at 2:13 PM, Lothar baier <Lothar@...> wrote:

?I just looked it up and your 8753 needs FW 7.68 or higher for the ECAL to work as well as the 85097B PC interface , only 85090 series ECAL are supported


Re: HP 8662A Keycap "BACK SPACE" needed

 

Hi Peter,
you got mail, please check your spam folder...
Regards
Chris


Re: HP 8594E upgrade CRT to LCD

 

Stuart not to throw stones. But I have seen a fair number of tek lcd displays that have gone to heck. Normally something like crystal leakage. (I just made that up). I can think of things like the tek TDR 1503 non crt versions yet the crt versions still work.
Just saying we hobbiest users use the gear far past any expectation that it should work.
Regards
Paul
WB8TSL


Re: eCal & HP-8753E

Lothar baier
 

开云体育

According to the compatibility matrix ONLY 85090 series ECAL are supported by the 8753 series analyzers , the N469X kits can be used however with the 8719/20/22 series but in any instance the 85097B is needed and the VNA must have the appropriate FW revision

?

Sent from for Windows

?

From: Steve Draper via groups.io
Sent: Wednesday, February 23, 2022 1:22 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] eCal & HP-8753E

?

The 8753e should also work with some newer ECAL units(such as the n4691a/b) but finding the unique cable they require is tough. I was not able to locate one when I was searching.

?


Re: eCal & HP-8753E

 

The 8753e should also work with some newer ECAL units(such as the n4691a/b) but finding the unique cable they require is tough. I was not able to locate one when I was searching.


Re: eCal & HP-8753E

Lothar baier
 

the E will work just as well as the ES/ET and even the D will work but the D only works if you have the flash based CPU board , in any case the analyzer must have FW 7.68 or higher


Re: eCal & HP-8753E

Lothar baier
 

I just looked it up and your 8753 needs FW 7.68 or higher for the ECAL to work as well as the 85097B PC interface , only 85090 series ECAL are supported


Re: eCal & HP-8753E

Lothar baier
 

no , the older VNA required the use of either a 85060C controller or a 85097A PC interface ,? there were also some FW requirements on the 8753 as far as i remember?
on a side note i want to caution you on ECAL kits , having used ECAL on both Rohde and Agilent units i have to warn you that ECAL units are very easy to damage and extremly sensitive to ESD?


looking for HP 8340/41B Manual change supplement for 2804A and 2812A Serial# range

Lothar baier
 

i have a 8341B that has SN 2804A however my 8510C testset requires a source with SN 2812A so i am trying to figure out what changes so i can possibly upgrade my 8341B?


Re: massive attachments

 

On Wed, Feb 23, 2022 at 12:33 PM, Roy J. Tellason, Sr. wrote:
I'd appreciate it if you guys could simply put such large files where somebody can get at them, and provide a link, for those that want it, rather than forcing those of us who get things this way to download for no immediate benefit.

Hi Roy,

In your subscription to this group, go to Advanced and you will see the maximum attachment size option, below:


Max Attachment Size

?? ?? ?? ?? ? Unlimited ?? ?? ?? ?? ? 0 (no attachments) ?? ?? ?? ?? ? 100KB ?? ?? ?? ?? ? 500KB ?? ?? ?? ?? ? 5MB ? ? ?? ?? ?

Maximum size of attachments sent to you in emails. For attachments larger than this size, links to retrieve them will be provided instead.


You can control the attachments which are sent to you.

Also, you can change Digest or Summary rather than All Messages, which allows you to choose individual messages.


Take care,

Larry


Re: eCal & HP-8753E

 

I'd be interested in the power supply if it is available.

Cheers!

Bruce


Quoting Steve Draper <steve@...>:

Get an 85097 as others have mentioned. An ECAL is totally worth it. I used
them with my 8753e when I had it. I actually still have the 85097 and
cables if you're interested.