OK, I'll bite--

What kind of power supply would you recommend?? I always thought that a regulated supply would be the safest.?

Were these linear or switcher supplies that produced the spikes?

Thanks,

Dave


On 8/20/2018 2:31 PM, Alan Melia via Groups.Io wrote:

During 10 years or so doing life tests on transistors in the 1960s and 70s I would not use regulated power supplies for powering the devices during their life test. When asked, I demonstrated the switch on pulse with a Tek storage scope. ok it was only microseconds wide but peaked at the unregulated level.~35v .....enough to damage 24v transistors.This can also happen with a transcient load. It is to do with the bandwidth of the feedback loop.
?
Alan
G3NYK


Joel Kist, was Re: [HP-Agilent-Keysight-equipment] HP 6012B volt zero and current regulation

 

On 08/20/2018 02:48 PM, Mark Wendt wrote:
Off subject – I still get duplicates – Joel?

By chance are you subscribed to the group with two different email
addresses? Check the headers of the emails that are dupes the next time
you receive them.?
Yes, he is. There have been no duplicate messages here. That's a
Yahoo problem specifically.

Joel, you are subscribed at both your sbcglobal.net address and your
att.net address. If you let me know which one you want to keep, I can
chop the other one for you.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: A huge thank you!

kevin kearns
 

开云体育

Well done David!

----- Original Message -----
Sent: Monday, August 20, 2018 8:49 PM
Subject: Re: [HP-Agilent-Keysight-equipment] A huge thank you!

Awesome move... thanks to all involved in it! Bravo!

Phil, N6OMM?

On Aug 20, 2018, at 7:08 AM, James Allen <allen3500@...> wrote:

Ditto from me Dave!!!

Jim


Re: A huge thank you!

 

开云体育

Awesome move... thanks to all involved in it! Bravo!

Phil, N6OMM?

On Aug 20, 2018, at 7:08 AM, James Allen <allen3500@...> wrote:

Ditto from me Dave!!!

Jim


Re: HP 6012B volt zero and current regulation

 

开云体育

I did a lot of power supply testing for HP, Agilent, and NGC, including the 6012A/B’s,
Here is the spec sheet:?. ?What you want to look at is the transient response - right hand column. ?Calibration transient response is tested for both voltage and current modes and is tested full to no resistive load, and also with 100/90% resistive load change. ?2ms/100mv. ?I believe that a min. 5 min. warmup time is required for specs to apply.
Don Bitters


Re: Great move

 

Also thank you for the move. I did not realize it until I got a redirect and log in request.
I also didn’t realize I had not been getting the summaries except rarely.
Nicely done. I now have the website referenced and logged.
Thank you,
Don Bitters (ret. HP/Agilent/NGC engineer)


Re: 8566B SA all PLLs unlocked

 

I too have seen a failed power resistor. ?Other failed item maybe be a leaky diode that then shuts down the power supply. ?If the -40v supply is other than 0, -40v suspect a leaky diode, or shorted or leaky cap on the affected supply. ?
You will also get a similar symptom if you have an ext 10 MHz ref and it drops out, or you lose the internal 10MHz, or the 100MHz loop is down.

Don Bitters


Re: HP 6012B volt zero and current regulation

 



On Mon, Aug 20, 2018, 14:22 Joel R Kist <joel-kist@...> wrote:

Off subject – I still get duplicates – Joel?


By chance are you subscribed to the group with two different email addresses? Check the headers of the emails that are dupes the next time you receive them.?

Mark


Re: HP 6012B volt zero and current regulation

 

?
During 10 years or so doing life tests on transistors in the 1960s and 70s I would not use regulated power supplies for powering the devices during their life test. When asked, I demonstrated the switch on pulse with a Tek storage scope. ok it was only microseconds wide but peaked at the unregulated level.~35v .....enough to damage 24v transistors.This can also happen with a transcient load. It is to do with the bandwidth of the feedback loop.
?
Alan
G3NYK
?

----- Original Message -----
Sent: Monday, August 20, 2018 7:18 PM
Subject: Re: [HP-Agilent-Keysight-equipment] HP 6012B volt zero and current regulation

Hi

In my experience most bench power supplies show a similar behaviour. ?At one time I suffered repeated damage to equipment that was powered from a bench supply, when checked with a storage scope it went to a high voltage on turn on before settling back to the set voltage.

I still think that for valid measurements you cannot use a light bulb, they are NOT a linear devices if you do an applied voltage plot against observed resistance, from a sixtey year old memory of an experiment they are square law devices

73 George G6HIG ??



On Monday, August 20, 2018 5:50 PM, "wilson2115@..." <wilson2115@...> wrote:


I refined my test...
Instead of ramping the voltage with the knob. I basically set the voltage to 20V and the amperage to 2 amps. I connected the voltmeter in series with the bulb and performed a dynamic load test by opening and closing the circuit by touching the probe lead of the voltmeter to the output of the power supply... I also have the Fluke voltmeter set to peak amperage measurement and these are the results that I found.

With the 6012B set to 20V and 2A, upon closing the circuit the peak amps are 4 amps but quickly settles to 2 amps
With the 6024A set to 20V and 2 A, upon closing the circuit the peaks amps are about 2.5 but quickly settles to 2 amps.

The important observation here is the 6012B settles back to 2 amps, it seems the settle time is longer and the deviation is larger than the 6024A. Although these are two completely different supplies. If anything I could be nit picking here on the 6012B for its regulation settle time and overshoot, which is why I wanted to see if someone else had a similar supply and could see if they could duplicate the same results to see if this is normal...

Thanks for the help



Re: HP 6012B volt zero and current regulation

 

开云体育

Also many bench supplies have very large output capacitors. I have a smaller HP on my bench and recorded over 100 amps instantaneous on a short with a couple feet of #10 leads. That was all from the output cap, it’s a 10 amp supply.?


Peter

On Aug 20, 2018, at 2:18 PM, george edmonds via Groups.Io <G6HIG@...> wrote:

Hi

In my experience most bench power supplies show a similar behaviour. ?At one time I suffered repeated damage to equipment that was powered from a bench supply, when checked with a storage scope it went to a high voltage on turn on before settling back to the set voltage.

I still think that for valid measurements you cannot use a light bulb, they are NOT a linear devices if you do an applied voltage plot against observed resistance, from a sixtey year old memory of an experiment they are square law devices

73 George G6HIG ??



On Monday, August 20, 2018 5:50 PM, "wilson2115@..." <wilson2115@...> wrote:


I refined my test...
Instead of ramping the voltage with the knob. I basically set the voltage to 20V and the amperage to 2 amps. I connected the voltmeter in series with the bulb and performed a dynamic load test by opening and closing the circuit by touching the probe lead of the voltmeter to the output of the power supply... I also have the Fluke voltmeter set to peak amperage measurement and these are the results that I found.

With the 6012B set to 20V and 2A, upon closing the circuit the peak amps are 4 amps but quickly settles to 2 amps
With the 6024A set to 20V and 2 A, upon closing the circuit the peaks amps are about 2.5 but quickly settles to 2 amps.

The important observation here is the 6012B settles back to 2 amps, it seems the settle time is longer and the deviation is larger than the 6024A. Although these are two completely different supplies. If anything I could be nit picking here on the 6012B for its regulation settle time and overshoot, which is why I wanted to see if someone else had a similar supply and could see if they could duplicate the same results to see if this is normal...

Thanks for the help



Re: HP 6012B volt zero and current regulation

 

开云体育

Off subject – I still get duplicates – Joel STS SS

?

From: [email protected] [mailto:[email protected]] On Behalf Of wilson2115@...
Sent: Monday, August 20, 2018 12:50 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 6012B volt zero and current regulation

?

I refined my test...
Instead of ramping the voltage with the knob. I basically set the voltage to 20V and the amperage to 2 amps. I connected the voltmeter in series with the bulb and performed a dynamic load test by opening and closing the circuit by touching the probe lead of the voltmeter to the output of the power supply... I also have the Fluke voltmeter set to peak amperage measurement and these are the results that I found.

With the 6012B set to 20V and 2A, upon closing the circuit the peak amps are 4 amps but quickly settles to 2 amps
With the 6024A set to 20V and 2 A, upon closing the circuit the peaks amps are about 2.5 but quickly settles to 2 amps.

The important observation here is the 6012B settles back to 2 amps, it seems the settle time is longer and the deviation is larger than the 6024A. Although these are two completely different supplies If anything I could be nit picking here on the 6012B for its regulation settle time and overshoot, which is why I wanted to see if someone else had a similar supply and could see if they could duplicate the same results to see if this is normal...

Thanks for the help


Re: HP 6012B volt zero and current regulation

 

Hi

In my experience most bench power supplies show a similar behaviour. ?At one time I suffered repeated damage to equipment that was powered from a bench supply, when checked with a storage scope it went to a high voltage on turn on before settling back to the set voltage.

I still think that for valid measurements you cannot use a light bulb, they are NOT a linear devices if you do an applied voltage plot against observed resistance, from a sixtey year old memory of an experiment they are square law devices

73 George G6HIG ??



On Monday, August 20, 2018 5:50 PM, "wilson2115@..." <wilson2115@...> wrote:


I refined my test...
Instead of ramping the voltage with the knob. I basically set the voltage to 20V and the amperage to 2 amps. I connected the voltmeter in series with the bulb and performed a dynamic load test by opening and closing the circuit by touching the probe lead of the voltmeter to the output of the power supply... I also have the Fluke voltmeter set to peak amperage measurement and these are the results that I found.

With the 6012B set to 20V and 2A, upon closing the circuit the peak amps are 4 amps but quickly settles to 2 amps
With the 6024A set to 20V and 2 A, upon closing the circuit the peaks amps are about 2.5 but quickly settles to 2 amps.

The important observation here is the 6012B settles back to 2 amps, it seems the settle time is longer and the deviation is larger than the 6024A. Although these are two completely different supplies. If anything I could be nit picking here on the 6012B for its regulation settle time and overshoot, which is why I wanted to see if someone else had a similar supply and could see if they could duplicate the same results to see if this is normal...

Thanks for the help



Re: HP 6012B volt zero and current regulation

 

I refined my test...
Instead of ramping the voltage with the knob. I basically set the voltage to 20V and the amperage to 2 amps. I connected the voltmeter in series with the bulb and performed a dynamic load test by opening and closing the circuit by touching the probe lead of the voltmeter to the output of the power supply... I also have the Fluke voltmeter set to peak amperage measurement and these are the results that I found.

With the 6012B set to 20V and 2A, upon closing the circuit the peak amps are 4 amps but quickly settles to 2 amps
With the 6024A set to 20V and 2 A, upon closing the circuit the peaks amps are about 2.5 but quickly settles to 2 amps.

The important observation here is the 6012B settles back to 2 amps, it seems the settle time is longer and the deviation is larger than the 6024A. Although these are two completely different supplies. If anything I could be nit picking here on the 6012B for its regulation settle time and overshoot, which is why I wanted to see if someone else had a similar supply and could see if they could duplicate the same results to see if this is normal...

Thanks for the help


Great move

 

Sincerely,

James L. Hudson
WA5JAT


Re: 8566B SA all PLLs unlocked

 

Hello Calvin,
did you check the 6 power resistors? It did happen to me that a broken one on A1A6 once killed an opamp in the IF/display unit.
best regards,
Peter


so glad to be here!

walter shawlee
 

many thanks to those that moved the group, I hope our new home proves to be a good one.
all the best,

walter (walter2 -at- sphere.bc.ca
sphere research corp.


Re: A huge thank you!

 

Thanks for the smooth move!


On Mon, Aug 20, 2018 at 6:32 AM Keith Monahan via Groups.Io <keith=[email protected]> wrote:
Thanks indeed!

I guess there was no problem at all transferring messages?

I scraped the group awhile ago when I thought yahoo might disappear. No
jails, no throttling, no VPN required. It just worked. I think it was a
little slow, maybe overnight for 88K messages.

Glad to see you had no troubles!

Thanks again!
Keith

On 8/20/2018 9:08 AM, James Allen wrote:
> Ditto from me Dave!!!
>
> Jim
>




Re: migration test message

 

On Mon, 20 Aug 2018, f1chf wrote:

Loud and clear :)

hell just to check !
thanks
---
*
* KSI@home KOI8 Net < > The impossible we do immediately. *
* Las Vegas NV, USA < > Miracles require 24-hour notice. *
*


Re: Welcome to [email protected]

 

very nice and smooth :-)
congratulations!


Re: A huge thank you!

 

Thanks indeed!

I guess there was no problem at all transferring messages?

I scraped the group awhile ago when I thought yahoo might disappear. No jails, no throttling, no VPN required. It just worked. I think it was a little slow, maybe overnight for 88K messages.

Glad to see you had no troubles!

Thanks again!
Keith

On 8/20/2018 9:08 AM, James Allen wrote:
Ditto from me Dave!!!
Jim