Alan, Here¡¯s how to set up the Impex SmartCAM to readout in pCi/ml and Flow in LPM
Record your original settings before changing things. Then open the Monitor tab then Instrument options. Check only Bypass Filter Log on and calibrate in cpm In the displayed column I¡¯ve moved these to display Temp Press Flow Alpha chronic Alpha acute Radon chronic Radon Acute Po-214 chronic Po-214 acute Po-212 chronic Po-212 acute Po-218 acute P-218 chronic When running I usually have the Top display show Radon Acute and the bottom display Radon Chronic. You¡¯ll need to multiply the pCi/ml x 1000 to get pCi/l. The EPA action level is 4 pCi/l at which point they think you¡¯ll die or go broke fixing it
On the Channel tab: Set display units pCi ( should already be there) I¡¯m using Long as 140 and short as 60 sec. it was 240 and 120 but I¡¯m looking for quick response. Isotope window should be Radon Chronic ¨C you can¡¯t change it. Change the half life from E-18 to 186 sec which is the Po-218 half life that the radon will be calc¡¯ed from. That will get rid of the E-19 extension Check Archive data, Show differential, and show activity. That will get rid of the Radon pCih/l displayed and put radon activity pCi in its place on the screen. Equilibrium factor is how much Po-218 has platted out on the walls. I have generally used 0.3 but they are using the industry standard of 0.5. leave it 0.5
The other tabs ( Po-212 chronic etc. ) I haven¡¯t played with yet but the half lives and time may have to be adjusted there as well and I¡¯ll fool with that later.
If you go to the View tab and check Raw data Channel it will display raw cps. I leave that unchecked unless I¡¯m doing a manual radon calc and need those numbers When shutting down under the File tab use the Shut Down option. OK lets report those radon room numbers. A this point don¡¯t sniff any sources as you only have one filter to play with and you¡¯ll be surprised to see what¡¯s left showing on the Spectra the next morning and just from room air.
To change flow from CFM to LPM readout Go to the channels tab then to last tab Flow Change the display units to l/m, Under the Alarms box check first box change set boxes to ABOVE, 49.5, High Alarm ¨C leave all latch boxes unchecked check second box, set box to BELOW, 19.8, and Low Limit. check third box, set box to BELOW, 15, Low Fail Dud
toggle quoted message
Show quoted text
-----Original Message----- From: [email protected] [mailto: [email protected]] On Behalf Of Geo Dowell Sent: Saturday, June 11, 2022 5:29 AM To: geoelectronics at rallstech.net; AlphaBetaSpectrometry Cc: Alan Jose Subject: Re: Alpha and Beta particle nuclear decay studies. This Weekend From: Alan Jose <Alan.Jose@...> To: geoelectronics@... Sent: Fri, 10 Jun 2022 22:07:01 -0400 (EDT) Subject: This Weekend "This weekend I plan on getting in plenty of lab time. We¡¯ll see. Alan" Good Alan. If you set up the IMPEX, and want to remove the filter card, be sure to ask me or read the instructions how th release it by rotating that lever- pull out the locking pin, rotate that lever as far as it will go, that releases the clamping mechanism so the filter can go in and out without damage. I messed up my filter before understanding that, but already have new filters coming from China. George
|
Arrrrrrgh!? It's over 100 in the shop and I misread your instructions and messed up the alarms under Stack Flow.? Can someone give me the default settings for those??
And the screen calibration instructions,? can't seem to find those right now.?
Thanks.?
toggle quoted message
Show quoted text
On Sat, Jun 11, 2022, 2:33 PM Dude < dfemer@...> wrote: Alan,
Here¡¯s how to set up the Impex SmartCAM to readout in pCi/ml and Flow in LPM
Record your original settings before changing things.
Then open the? Monitor tab then Instrument options. Check only Bypass Filter Log on and calibrate in cpm
In the displayed column I¡¯ve moved these to display
Temp
Press
?Flow
Alpha chronic
Alpha acute
Radon chronic
Radon Acute
Po-214 chronic
Po-214 acute
Po-212 chronic
Po-212 acute
Po-218 acute
P-218 chronic
When running I usually have the Top display show Radon Acute and the bottom display Radon Chronic.? You¡¯ll need to multiply the pCi/ml x 1000 to get pCi/l.? The EPA action level is 4 pCi/l at which point they think you¡¯ll die or go broke fixing it
On the Channel tab:
Set display units pCi ( should already be there)
I¡¯m using Long as 140 and short as 60 sec. it was 240 and 120 but I¡¯m looking for quick response.
Isotope window should be Radon Chronic ¨C you can¡¯t change it.
Change the half life from E-18 to 186 sec which is the Po-218 half life that the radon will be calc¡¯ed from. That will get rid of the E-19 extension
Check Archive data, Show differential, and show activity. That will get rid of the Radon pCih/l displayed and put radon activity pCi in its place on the screen.
Equilibrium factor is how much Po-218 has platted out on the walls. I have generally used 0.3 but they are using the industry standard of 0.5. leave it 0.5
The other tabs ( Po-212 chronic etc. ) I haven¡¯t played with yet? but the half lives and time may have to be adjusted there as well and I¡¯ll fool with that later.
If you go to the View tab and check Raw data Channel it will display raw cps. I leave that unchecked unless I¡¯m doing a manual radon calc and need those numbers
When shutting down under the File tab use the Shut Down option.
OK lets report those radon room numbers. A this point don¡¯t sniff any sources as you only have one filter to play with and you¡¯ll be surprised to see what¡¯s left showing on the Spectra the next morning and just from room air.
To change flow from CFM to LPM readout
Go to the channels tab then to last tab Flow
Change the display units to l/m, Under the Alarms box
check first box change set boxes to ABOVE, 49.5, High Alarm ¨C leave all latch boxes unchecked
check second box, set box to BELOW, 19.8, and? Low Limit.
check third box, set box to BELOW, 15, Low Fail
Dud
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Geo Dowell
Sent: Saturday, June 11, 2022 5:29 AM
To: geoelectronics at ; AlphaBetaSpectrometry
Cc: Alan Jose
Subject: Re: Alpha and Beta particle nuclear decay studies. This Weekend
From: Alan Jose <Alan.Jose@...>
To: geoelectronics@...
Sent: Fri, 10 Jun 2022 22:07:01 -0400 (EDT)
Subject: This Weekend
"This weekend I plan on getting in plenty of lab time. We¡¯ll see.
Alan"
Good Alan. If you set up the IMPEX, and want to remove the filter card, be sure to ask me or read the instructions how th release it by rotating that lever- pull out the locking pin, rotate that lever as far as it will go, that releases the clamping mechanism so the filter can go in and out without damage. I messed up my filter before understanding that, but already have new filters coming from China.
George
|
Not running for too long,? just while changing settings.? So for has been open most of the day,? with small exhaust fan running too.
toggle quoted message
Show quoted text
On Sat, Jun 11, 2022, 2:33 PM Dude < dfemer@...> wrote: Alan,
Here¡¯s how to set up the Impex SmartCAM to readout in pCi/ml and Flow in LPM
Record your original settings before changing things.
Then open the? Monitor tab then Instrument options. Check only Bypass Filter Log on and calibrate in cpm
In the displayed column I¡¯ve moved these to display
Temp
Press
?Flow
Alpha chronic
Alpha acute
Radon chronic
Radon Acute
Po-214 chronic
Po-214 acute
Po-212 chronic
Po-212 acute
Po-218 acute
P-218 chronic
When running I usually have the Top display show Radon Acute and the bottom display Radon Chronic.? You¡¯ll need to multiply the pCi/ml x 1000 to get pCi/l.? The EPA action level is 4 pCi/l at which point they think you¡¯ll die or go broke fixing it
On the Channel tab:
Set display units pCi ( should already be there)
I¡¯m using Long as 140 and short as 60 sec. it was 240 and 120 but I¡¯m looking for quick response.
Isotope window should be Radon Chronic ¨C you can¡¯t change it.
Change the half life from E-18 to 186 sec which is the Po-218 half life that the radon will be calc¡¯ed from. That will get rid of the E-19 extension
Check Archive data, Show differential, and show activity. That will get rid of the Radon pCih/l displayed and put radon activity pCi in its place on the screen.
Equilibrium factor is how much Po-218 has platted out on the walls. I have generally used 0.3 but they are using the industry standard of 0.5. leave it 0.5
The other tabs ( Po-212 chronic etc. ) I haven¡¯t played with yet? but the half lives and time may have to be adjusted there as well and I¡¯ll fool with that later.
If you go to the View tab and check Raw data Channel it will display raw cps. I leave that unchecked unless I¡¯m doing a manual radon calc and need those numbers
When shutting down under the File tab use the Shut Down option.
OK lets report those radon room numbers. A this point don¡¯t sniff any sources as you only have one filter to play with and you¡¯ll be surprised to see what¡¯s left showing on the Spectra the next morning and just from room air.
To change flow from CFM to LPM readout
Go to the channels tab then to last tab Flow
Change the display units to l/m, Under the Alarms box
check first box change set boxes to ABOVE, 49.5, High Alarm ¨C leave all latch boxes unchecked
check second box, set box to BELOW, 19.8, and? Low Limit.
check third box, set box to BELOW, 15, Low Fail
Dud
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Geo Dowell
Sent: Saturday, June 11, 2022 5:29 AM
To: geoelectronics at ; AlphaBetaSpectrometry
Cc: Alan Jose
Subject: Re: Alpha and Beta particle nuclear decay studies. This Weekend
From: Alan Jose <Alan.Jose@...>
To: geoelectronics@...
Sent: Fri, 10 Jun 2022 22:07:01 -0400 (EDT)
Subject: This Weekend
"This weekend I plan on getting in plenty of lab time. We¡¯ll see.
Alan"
Good Alan. If you set up the IMPEX, and want to remove the filter card, be sure to ask me or read the instructions how th release it by rotating that lever- pull out the locking pin, rotate that lever as far as it will go, that releases the clamping mechanism so the filter can go in and out without damage. I messed up my filter before understanding that, but already have new filters coming from China.
George
|
And not long after that.? That spectrum really starts to build up.?
toggle quoted message
Show quoted text
Not running for too long,? just while changing settings.? So for has been open most of the day,? with small exhaust fan running too.
On Sat, Jun 11, 2022, 2:33 PM Dude < dfemer@...> wrote: Alan,
Here¡¯s how to set up the Impex SmartCAM to readout in pCi/ml and Flow in LPM
Record your original settings before changing things.
Then open the? Monitor tab then Instrument options. Check only Bypass Filter Log on and calibrate in cpm
In the displayed column I¡¯ve moved these to display
Temp
Press
?Flow
Alpha chronic
Alpha acute
Radon chronic
Radon Acute
Po-214 chronic
Po-214 acute
Po-212 chronic
Po-212 acute
Po-218 acute
P-218 chronic
When running I usually have the Top display show Radon Acute and the bottom display Radon Chronic.? You¡¯ll need to multiply the pCi/ml x 1000 to get pCi/l.? The EPA action level is 4 pCi/l at which point they think you¡¯ll die or go broke fixing it
On the Channel tab:
Set display units pCi ( should already be there)
I¡¯m using Long as 140 and short as 60 sec. it was 240 and 120 but I¡¯m looking for quick response.
Isotope window should be Radon Chronic ¨C you can¡¯t change it.
Change the half life from E-18 to 186 sec which is the Po-218 half life that the radon will be calc¡¯ed from. That will get rid of the E-19 extension
Check Archive data, Show differential, and show activity. That will get rid of the Radon pCih/l displayed and put radon activity pCi in its place on the screen.
Equilibrium factor is how much Po-218 has platted out on the walls. I have generally used 0.3 but they are using the industry standard of 0.5. leave it 0.5
The other tabs ( Po-212 chronic etc. ) I haven¡¯t played with yet? but the half lives and time may have to be adjusted there as well and I¡¯ll fool with that later.
If you go to the View tab and check Raw data Channel it will display raw cps. I leave that unchecked unless I¡¯m doing a manual radon calc and need those numbers
When shutting down under the File tab use the Shut Down option.
OK lets report those radon room numbers. A this point don¡¯t sniff any sources as you only have one filter to play with and you¡¯ll be surprised to see what¡¯s left showing on the Spectra the next morning and just from room air.
To change flow from CFM to LPM readout
Go to the channels tab then to last tab Flow
Change the display units to l/m, Under the Alarms box
check first box change set boxes to ABOVE, 49.5, High Alarm ¨C leave all latch boxes unchecked
check second box, set box to BELOW, 19.8, and? Low Limit.
check third box, set box to BELOW, 15, Low Fail
Dud
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Geo Dowell
Sent: Saturday, June 11, 2022 5:29 AM
To: geoelectronics at ; AlphaBetaSpectrometry
Cc: Alan Jose
Subject: Re: Alpha and Beta particle nuclear decay studies. This Weekend
From: Alan Jose <Alan.Jose@...>
To: geoelectronics@...
Sent: Fri, 10 Jun 2022 22:07:01 -0400 (EDT)
Subject: This Weekend
"This weekend I plan on getting in plenty of lab time. We¡¯ll see.
Alan"
Good Alan. If you set up the IMPEX, and want to remove the filter card, be sure to ask me or read the instructions how th release it by rotating that lever- pull out the locking pin, rotate that lever as far as it will go, that releases the clamping mechanism so the filter can go in and out without damage. I messed up my filter before understanding that, but already have new filters coming from China.
George
|
Nick we¡¯re not using stack
flow just turn it off. That is for a remote air flow sensor in the Stack.
Screen Calibration
instructions are in the manual on page 12-9. You¡¯ll need a touch pen with a small
tip to get it accurate.
Dud
?
?
Arrrrrrgh!? It's over 100 in the shop and I misread your
instructions and messed up the alarms under Stack Flow.? Can someone give
me the default settings for those??
And the screen calibration instructions,? can't seem to find those
right now.?
?
toggle quoted message
Show quoted text
On Sat, Jun 11, 2022, 2:33 PM Dude <dfemer@...>
wrote:
Alan,
Here¡¯s how to set up the Impex SmartCAM to readout in pCi/ml and Flow in LPM
Record your original settings before changing things.
Then open the? Monitor tab then Instrument options. Check only Bypass
Filter Log on and calibrate in cpm
In the displayed column I¡¯ve moved these to display
Temp
Press
?Flow
Alpha chronic
Alpha acute
Radon chronic
Radon Acute
Po-214 chronic
Po-214 acute
Po-212 chronic
Po-212 acute
Po-218 acute
P-218 chronic
When running I usually have the Top display show Radon Acute and the bottom
display Radon Chronic.? You¡¯ll need to multiply the pCi/ml x 1000 to get
pCi/l.? The EPA action level is 4 pCi/l at which point they think you¡¯ll
die or go broke fixing it
On the Channel tab:
Set display units pCi ( should already be there)
I¡¯m using Long as 140 and short as 60 sec. it was 240 and 120 but I¡¯m looking
for quick response.
Isotope window should be Radon Chronic ¨C you can¡¯t change it.
Change the half life from E-18 to 186 sec which is the Po-218 half life that
the radon will be calc¡¯ed from. That will get rid of the E-19 extension
Check Archive data, Show differential, and show activity. That will get rid of
the Radon pCih/l displayed and put radon activity pCi in its place on the
screen.
Equilibrium factor is how much Po-218 has platted out on the walls. I have
generally used 0.3 but they are using the industry standard of 0.5. leave it 0.5
The other tabs ( Po-212 chronic etc. ) I haven¡¯t played with yet? but the
half lives and time may have to be adjusted there as well and I¡¯ll fool with
that later.
If you go to the View tab and check Raw data Channel it will display raw cps. I
leave that unchecked unless I¡¯m doing a manual radon calc and need those
numbers
When shutting down under the File tab use the Shut Down option.
OK lets report those radon room numbers. A this point don¡¯t sniff any sources
as you only have one filter to play with and you¡¯ll be surprised to see what¡¯s
left showing on the Spectra the next morning and just from room air.
To change flow from CFM to LPM readout
Go to the channels tab then to last tab Flow
Change the display units to l/m, Under the Alarms box
check first box change set boxes to ABOVE, 49.5, High Alarm ¨C leave all latch
boxes unchecked
check second box, set box to BELOW, 19.8, and? Low Limit.
check third box, set box to BELOW, 15, Low Fail
Dud
-----Original Message-----
From: [email protected]
[mailto:[email protected]]
On Behalf Of Geo Dowell
Sent: Saturday, June 11, 2022 5:29 AM
To: geoelectronics at ;
AlphaBetaSpectrometry
Cc: Alan Jose
Subject: Re: Alpha and Beta particle nuclear decay studies. This Weekend
From: Alan Jose <Alan.Jose@...>
To: geoelectronics@...
Sent: Fri, 10 Jun 2022 22:07:01 -0400 (EDT)
Subject: This Weekend
"This weekend I plan on getting in plenty of lab time. We¡¯ll see.
Alan"
Good Alan. If you set up the IMPEX, and want to remove the filter card, be sure
to ask me or read the instructions how th release it by rotating that lever-
pull out the locking pin, rotate that lever as far as it will go, that releases
the clamping mechanism so the filter can go in and out without damage. I messed
up my filter before understanding that, but already have new filters coming
from China.
George
|
Nick,
That looks? like outside
air for radon or its only been on a couple of minutes. The unit needs to run
about 60 minutes to get a fairly acute reading but getting pretty close should be
about 20mins with at least some Rn activity.
Dud
?
?
And not long after that.? That spectrum really starts to build
up.?
?
toggle quoted message
Show quoted text
On Sun, Jun 12, 2022, 7:52 PM Nick Andrews <nickjandrews@...> wrote:
Not running for too long,? just while changing settings.? So
for has been open most of the day,? with small exhaust fan running too.
?
On Sat, Jun 11, 2022, 2:33 PM Dude <dfemer@...> wrote:
Alan,
Here¡¯s how to set up the Impex SmartCAM to readout in pCi/ml and Flow in LPM
Record your original settings before changing things.
Then open the? Monitor tab then Instrument options. Check only Bypass
Filter Log on and calibrate in cpm
In the displayed column I¡¯ve moved these to display
Temp
Press
?Flow
Alpha chronic
Alpha acute
Radon chronic
Radon Acute
Po-214 chronic
Po-214 acute
Po-212 chronic
Po-212 acute
Po-218 acute
P-218 chronic
When running I usually have the Top display show Radon Acute and the bottom
display Radon Chronic.? You¡¯ll need to multiply the pCi/ml x 1000 to get
pCi/l.? The EPA action level is 4 pCi/l at which point they think you¡¯ll
die or go broke fixing it
On the Channel tab:
Set display units pCi ( should already be there)
I¡¯m using Long as 140 and short as 60 sec. it was 240 and 120 but I¡¯m looking
for quick response.
Isotope window should be Radon Chronic ¨C you can¡¯t change it.
Change the half life from E-18 to 186 sec which is the Po-218 half life that
the radon will be calc¡¯ed from. That will get rid of the E-19 extension
Check Archive data, Show differential, and show activity. That will get rid of
the Radon pCih/l displayed and put radon activity pCi in its place on the
screen.
Equilibrium factor is how much Po-218 has platted out on the walls. I have
generally used 0.3 but they are using the industry standard of 0.5. leave it
0.5
The other tabs ( Po-212 chronic etc. ) I haven¡¯t played with yet? but the
half lives and time may have to be adjusted there as well and I¡¯ll fool with
that later.
If you go to the View tab and check Raw data Channel it will display raw cps. I
leave that unchecked unless I¡¯m doing a manual radon calc and need those
numbers
When shutting down under the File tab use the Shut Down option.
OK lets report those radon room numbers. A this point don¡¯t sniff any sources
as you only have one filter to play with and you¡¯ll be surprised to see what¡¯s
left showing on the Spectra the next morning and just from room air.
To change flow from CFM to LPM readout
Go to the channels tab then to last tab Flow
Change the display units to l/m, Under the Alarms box
check first box change set boxes to ABOVE, 49.5, High Alarm ¨C leave all latch
boxes unchecked
check second box, set box to BELOW, 19.8, and? Low Limit.
check third box, set box to BELOW, 15, Low Fail
Dud
-----Original Message-----
From: [email protected]
[mailto:[email protected]]
On Behalf Of Geo Dowell
Sent: Saturday, June 11, 2022 5:29 AM
To: geoelectronics at ;
AlphaBetaSpectrometry
Cc: Alan Jose
Subject: Re: Alpha and Beta particle nuclear decay studies. This Weekend
From: Alan Jose <Alan.Jose@...>
To: geoelectronics@...
Sent: Fri, 10 Jun 2022 22:07:01 -0400 (EDT)
Subject: This Weekend
"This weekend I plan on getting in plenty of lab time. We¡¯ll see.
Alan"
Good Alan. If you set up the IMPEX, and want to remove the filter card, be sure
to ask me or read the instructions how th release it by rotating that lever-
pull out the locking pin, rotate that lever as far as it will go, that releases
the clamping mechanism so the filter can go in and out without damage. I messed
up my filter before understanding that, but already have new filters coming
from China.
George
|
Yes, it was a short run.? It's also only a 20x20 (outside) steel building on a newish slab that varies from 7-14 inches thick, sitting pretty much directly on limestone in central TX.? It's not sealed tight and has a small 4" exhaust fan near the peak in one end running continuously.? And before I turned the unit on to change settings, the door had been open most of the day to run an air hose out so there was a lot of fresh air exchange.
I did go and just turn the stack flow alarms off.
toggle quoted message
Show quoted text
On Sun, Jun 12, 2022 at 8:10 PM Dude < dfemer@...> wrote:
Nick,
That looks? like outside
air for radon or its only been on a couple of minutes. The unit needs to run
about 60 minutes to get a fairly acute reading but getting pretty close should be
about 20mins with at least some Rn activity.
Dud
?
?
And not long after that.? That spectrum really starts to build
up.?
?
Not running for too long,? just while changing settings.? So
for has been open most of the day,? with small exhaust fan running too.
?
On Sat, Jun 11, 2022, 2:33 PM Dude <dfemer@...> wrote:
Alan,
Here¡¯s how to set up the Impex SmartCAM to readout in pCi/ml and Flow in LPM
Record your original settings before changing things.
Then open the? Monitor tab then Instrument options. Check only Bypass
Filter Log on and calibrate in cpm
In the displayed column I¡¯ve moved these to display
Temp
Press
?Flow
Alpha chronic
Alpha acute
Radon chronic
Radon Acute
Po-214 chronic
Po-214 acute
Po-212 chronic
Po-212 acute
Po-218 acute
P-218 chronic
When running I usually have the Top display show Radon Acute and the bottom
display Radon Chronic.? You¡¯ll need to multiply the pCi/ml x 1000 to get
pCi/l.? The EPA action level is 4 pCi/l at which point they think you¡¯ll
die or go broke fixing it
On the Channel tab:
Set display units pCi ( should already be there)
I¡¯m using Long as 140 and short as 60 sec. it was 240 and 120 but I¡¯m looking
for quick response.
Isotope window should be Radon Chronic ¨C you can¡¯t change it.
Change the half life from E-18 to 186 sec which is the Po-218 half life that
the radon will be calc¡¯ed from. That will get rid of the E-19 extension
Check Archive data, Show differential, and show activity. That will get rid of
the Radon pCih/l displayed and put radon activity pCi in its place on the
screen.
Equilibrium factor is how much Po-218 has platted out on the walls. I have
generally used 0.3 but they are using the industry standard of 0.5. leave it
0.5
The other tabs ( Po-212 chronic etc. ) I haven¡¯t played with yet? but the
half lives and time may have to be adjusted there as well and I¡¯ll fool with
that later.
If you go to the View tab and check Raw data Channel it will display raw cps. I
leave that unchecked unless I¡¯m doing a manual radon calc and need those
numbers
When shutting down under the File tab use the Shut Down option.
OK lets report those radon room numbers. A this point don¡¯t sniff any sources
as you only have one filter to play with and you¡¯ll be surprised to see what¡¯s
left showing on the Spectra the next morning and just from room air.
To change flow from CFM to LPM readout
Go to the channels tab then to last tab Flow
Change the display units to l/m, Under the Alarms box
check first box change set boxes to ABOVE, 49.5, High Alarm ¨C leave all latch
boxes unchecked
check second box, set box to BELOW, 19.8, and? Low Limit.
check third box, set box to BELOW, 15, Low Fail
Dud
-----Original Message-----
From: [email protected]
[mailto:[email protected]]
On Behalf Of Geo Dowell
Sent: Saturday, June 11, 2022 5:29 AM
To: geoelectronics at ;
AlphaBetaSpectrometry
Cc: Alan Jose
Subject: Re: Alpha and Beta particle nuclear decay studies. This Weekend
From: Alan Jose <Alan.Jose@...>
To: geoelectronics@...
Sent: Fri, 10 Jun 2022 22:07:01 -0400 (EDT)
Subject: This Weekend
"This weekend I plan on getting in plenty of lab time. We¡¯ll see.
Alan"
Good Alan. If you set up the IMPEX, and want to remove the filter card, be sure
to ask me or read the instructions how th release it by rotating that lever-
pull out the locking pin, rotate that lever as far as it will go, that releases
the clamping mechanism so the filter can go in and out without damage. I messed
up my filter before understanding that, but already have new filters coming
from China.
George
|